[time-nuts] Zeeman frequency oddness
paul swed
paulswedb at gmail.com
Mon Oct 29 20:05:17 UTC 2012
Corby
Thanks for the key part what to do when you need to fix the issue.
Regards
Paul
WB8TSL
On Sun, Oct 28, 2012 at 11:50 AM, <cdelect at juno.com> wrote:
> After the degauss you should recheck the Zeeman setting, which you did.
>
> You also verified the synthesizer frequency is set to the frequency that
> requires a C-field current high enough to reach 53.53.
>
> Since it is running at 48.21 you should increase the C-field setting
> until the central peak moves to 53.53.
>
> This is assuming that your unit has the resistor installed for 53.53.
>
> If not you will reach the end of the C-field pot before getting to 53.53!
>
> In that case you need to put a decade box (set to max resistance) across
> the C-field resistor and with the C-field set to 500 reduce the
> resistance until the central peak gets to 53.53.
>
> Then solder in a 1% fixed resistor (closest value to the decade box
> reading) across the C-field resistor.
>
> Now recheck and you will be able to use the C-field pot to reach exactly
> 53.53.
>
> Cheers!
>
> Corby Dawson
> ____________________________________________________________
> Woman is 53 But Looks 25
> Mom reveals 1 simple wrinkle trick that has angered doctors...
> http://thirdpartyoffers.juno.com/TGL3141/508d54a2aac4554a26ff4st02duc
>
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
More information about the Time-nuts_lists.febo.com
mailing list