[time-nuts] OCXO Vref (in)stability plot
Ed Palmer
ed_palmer at sasktel.net
Sun Apr 21 22:35:10 UTC 2013
I think that Vref is derived from the internal voltage regulator that
powers the oscillator. Is it possible that any noise will appear on
both and therefore, they'll cancel each other out? That could even
reduce power supply sensitivity.
You could test that by measuring ADEV with an EFC pot powered by Vref
vs. powered by an external quiet supply.
Ed
On 4/21/2013 2:47 PM, Poul-Henning Kamp wrote:
> Ok, this surprised me a LOT!
>
> I used my HP3458A to measure the stability of the Vref output from
> an IsoTemp OCXO-107 and the results were anything but stellar:
>
> The OCXO's Vref has 18 microVolt RMS noise
>
> For comparison my HP6626A power-supply has 2 microVolt RMS noise.
>
> Obviously, the Vref only has to be good enough to not be a relevant
> source of noise when used on the EFC input.
>
> If the OCXO has a EFC sensitivity of 10PPM/Volt, then a microVolt
> more or less means a 1e-11 change in frequency.
>
> The Vref is also quite sensitive to the OCXO supply voltage, but
> since it was powered by the HP6626A, that's not where the noise
> comes from.
More information about the Time-nuts_lists.febo.com
mailing list