Hi Rex,
>I don't think the rubidium module for these systems can be disciplined.
>I think they were only intended as a backup if the XO section or GPS
failed.
>The XO section does discipline but I don't know the details.
I was hoping since the rubidium module had a light for GPS and an input for
the GPS that it also did the same thing.
>Trying to trace out some of the circuits, I found that the EFC on the
>LPRO connector was not being used at all.
That is too bad, but that proves it all right.
I tried to feed a 1pps signal into the interface pin 9 to see what would
happen, but the NO GPS light never went off and it didn't pull the frequency
at all.
So far I've done a 4 hour run between the Rb and GPS and counted
139390000010 Rb cycles to 13939 GPS 1pps pulses or 10000000.0007174 Hz for
the Rb.