Tue Feb 19 12:35:52 UTC 2013


Hi

Keep in mind that the Tracor's came out before the Teflon coated bulbs. Without the Teflon, the Rb will combine with the glass of the bulb. The ones I have seen slowly turn a black. As they darken, the light transmission drops enough to stop the standard. Since it's a chemical reaction, there is no simple way to reverse it.

Bob
   
On Feb 19, 2013, at 1:36 AM, Stewart Cobb <stewart.cobb at gmail.com> wrote:

> Guys,
> 
> I'm repairing a 1960's vintage lab-grade rubidium standard, General
> Technology Corporation model 304-B.  Apparently Tracor bought GTC soon
> after this unit was made, because references to this as a "Tracor 304-B"
> seem to be more common.  I've made some progress, but now it seems like
> time to consult the hive mind.
> 
> The unit appears clean, but it doesn't lock.  I've read through old
> comments on the list regarding this unit, and I've downloaded a copy of the
> manual and schematics available at
> 
> <*http://sundry.i2phd.com/ServiceManual_304b.pdf>*
> 
> That file seems to contain a complete copy of the manual text, but some
> schematics are missing.  In particular, the schematics for the
> sweep/acquisition board (A8) and the three boards inside the physics
> package (the lamp oscillator (A13), the SRD driver (A12), and the photocell
> preamp (A11)) are not shown.  Does anyone know where to find copies of
> those schematics?
> 
> The main power supply voltage on my unit seems to have been deliberately
> adjusted lower than spec (18.54 V actual, versus 20 +/- 0.1V specified in
> the manual).  Replacing a resistor on the regulator board (that had smoked
> from overload due to the low voltage) didn't change the voltage much.  I
> had to crank the trimmer across half of its range to get the voltage back
> within spec.  Nothing in the regulator circuitry seemed to have drifted
> enough to change the setpoint that much.  Is there a reason why a tech
> would have deliberately set this voltage lower than spec, or did it just
> drift down over the years?
> 
> A frequency counter (GPSDO reference) shows that the crystal oven warms up
> as expected.  The output can be centered on 5 MHz and the sweep circuit
> covers a symmetrical range around 5 MHz as expected.  The ovens for the
> lamp and filter cell appear to warm up properly as well, judging from test
> points available on the A1 oven controller board.  The test point voltages
> don't quite match the ones in the PDF manual, but it looks like those
> readings were typed into each individual manual after being read off the
> particular unit that came with that manual.
> 
> The test point on the A5 board shows that 155 Hz resonance detector
> modulation is within spec.  The A6 filter-amplifier board test points show
> the system attempting (and failing) to detect 155 Hz and 310 Hz resonance
> signals coming back from the photocell.
> 
> The manual says that the A7 RF pre-driver board (the x14 multiplier) should
> be supplying 70 MHz at +13 dBm to the SRD driver inside the physics
> package.  That would be about 2.8Vpp, assuming a 50-ohm system.  Instead,
> it's supplying a clean 70 MHz at about 100mV into a 50-ohm load.  My best
> guess is that the final amplifier transistor on that board is blown,
> possibly from being operated with only a scope probe as a load (infinite
> VSWR).  Replacement transistors are on order.  Any other thoughts?
> 
> Obviously, the box won't lock until the RF input is the right level.  But
> it also requires the Rb lamp to light.  Corby Dawson posted to the list
> back on 12 November 2009:
> 
> "Tracor bulbs fail with a different mechanism and last maybe 10 years."
> 
> Anyone know what that "different" failure mechanism is?  Is it repairable
> in an ordinary lab, like the heat-gun trick for LPRO bulbs?  If not, is it
> feasible to build a "Frankenstein" replacement using something like an LPRO
> or FEI bulb?
> 
> Is it possible to tell whether the lamp is lit without opening the physics
> package?  If not, are there any tricks to opening the physics package?  Any
> precautions to take before doing so?
> 
> Any other comments on how to get this box working again?
> 
> Cheers!
> --Stu
> 
> Side note:  This unit was built during the era of "elastic seconds"
> (roughly, the 1960's).  It contains a board (A9) which digitally offsets
> the output frequency in increments of roughly 7E-10, without changing the
> rubidium resonance frequency or the C-field.  There's also a note in the
> manual saying that annual changes to the definition of the second may
> require replacing the rubidium resonance cell in the physics package with a
> new cell calibrated for the new second in the new year.  Leap seconds bring
> their own problems, but compared to dismantling your lab instruments every
> year, they're a breeze.