[time-nuts] Traceability after loss of LORAN and WWVB

Bob Camp lists at rtty.us
Sat Jun 1 23:39:44 UTC 2013


Hi

The cost of initially setting up your gear to run to LORAN was more than the cost today of setting up similar gear to GPS. If you were doing it right, the LORAN system was no more / or less free than the same thing with GPS.

Simply running a couple SRS LORAN's was not in it's day compliant with the requirements of "legal metrology" nor was it particularly accurate. Todays "run a GPSDO" is actually more accurate (but no more compliant) than your SRS LORAN.

Bob

On Jun 1, 2013, at 7:10 PM, Scott McGrath <scmcgrath at gmail.com> wrote:

> In the case the lab in question was accredited and we went through the audits of process and procedure and go through them to this day and we have the nice certs on the wall.
> 
> When LORAN went away we then had to use very expensive processes to MAINTAIN that traceability and accreditation which LORAN provided very inexpensively for many years.  
> 
> And it would be really nice to be able to replicate the low cost of LORAN based legal metrology that with what we have today which is GPS
> 
> We have a few GPSDO's but they are not connected as currently worthless for legal metrology we have a couple of 5071's which get shipped under power to another lab for legal traceability 
> 
> It was a lot easier and cheaper back when a couple of SRS LORAN units and our record keeping and a annual process audit was all that was required to maintain legal traceability
> 
> 
> 
> 
> 
> Sent from my iPhone
> 
> On Jun 1, 2013, at 6:19 PM, "Charles P. Steinmetz" <charles_steinmetz at lavabit.com> wrote:
> 
>> 
>> Bob wrote:
>> 
>>>> At least the way I read the pdf's NIST seems to believe that GPS is legally traceable to NIST. It is the same "measure and then look up the data" sort of thing that LORAN used to be.  Took a while to read through them all…
>> 
>> Yes, that is correct.
>> 
>> Magnus wrote:
>> 
>>> However, just taking time from GPS does not achieve NIST traceability.
>>>  *   *   *
>>> You can achieve NIST traceability (or to any other NIH) if you do a whole bunch of things _right_ and in accordance with relevant standards. Few do.
>> 
>> That is also correct.  Instruments are *not* "NIST traceable."  However, a measurement made with the equipment can be "NIST traceable" under certain conditions.  It is the PROCESS that produces traceability.  (I use "NIST" here for convenience -- any National Metrology Institute can be substituted.)
>> 
>> Sadly, you are also correct that few do the process right.  In my experience, about 80% of labs that claim they are calibrating frequency-measuring equipment to "NIST-traceable" standards really aren't.  (I've had this rather adversarial discussion with several dozen lab managers over the years.)  Of course, this does not mean their calibrations are not accurate, just that, from the standpoint of legal metrology, the instruments they calibrate cannot be used to make NIST-traceable measurements or perform NIST-traceable calibrations.  (When it comes to all of the "NIST calibrated" equipment you see on ebay, the figure approaches 100% very, very closely.)
>> 
>> The same is true the other way around.  You can have all of your equipment calibrated by a lab that works to NIST-traceable standards, but if you do not follow through with the traceability process the measurements you make with those instruments (including calibrations done using them) will not be NIST-traceable, from a legal metrology perspective.
>> 
>> Think of the traceability process as a chain.  One broken link and traceability vanishes.
>> 
>> One of the required criteria for traceability is *demonstrated competence*.  Generally, this is done by becoming accredited to the relevant ISO/IEC standard by an accreditation body.  No matter how sophisticated and meticulous someone is, if their lab is not accredited, they cannot make NIST traceable measurements or perform NIST-traceable calibrations.  Period.  This is where home labs and other "little guys" inevitably fail to preserve traceability.  Again, this does not mean that their measurements or calibrations are junk -- just that they are not NIST-traceable, as far as legal metrology is concerned.  The A2LA shows you what you need to do (see link in my last e-mail).
>> 
>> The OP may have thought he was making NIST-traceable measurements using WWVB and/or LORAN standards.  But if his lab was not accredited, that was not so.  Nothing has changed, in that regard, with the advent of GPSDOs (except that the uncertainty levels are better now).
>> 
>> Best regards,
>> 
>> Charles
>> 
>> 
>> 
>> 
>> _______________________________________________
>> time-nuts mailing list -- time-nuts at febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.




More information about the Time-nuts_lists.febo.com mailing list