[time-nuts] FTS 4060 Cesium Time & Frequency Standard, Datum
J. L. Trantham
jltran at att.net
Wed Oct 2 02:35:54 UTC 2013
Tom,
I don't have any FTS CS units but I would be suspicious that the 'ion pump
voltage' of '0' really means that the ion pump current is 0. And, if so, I
would be very suspicious of a dead HV supply or a shorted tube.
If it's anything like the HP units, you should be able to disconnect the CS
tube from the HV supply and measure the HV supply directly from the
connector on the chassis and, from the wire from the tube, measure the
resistance from the HV wire on the tube to ground.
On the HP units, HV should be about +3500 VDC and the resistance should be
infinite. If the HV is absent, it's a HV supply problem and if the
resistance shows a short, you might be able to 'open' it by using the HV
supply to charge a HV capacitor then connect that to the tube (be careful,
HV really can hurt you) and see if it might 'blow' open the short (I was
able to do that to 'resurrect' an HP tube).
Good luck.
Joe
-----Original Message-----
From: time-nuts-bounces at febo.com [mailto:time-nuts-bounces at febo.com] On
Behalf Of Tom Miller
Sent: Tuesday, October 01, 2013 8:53 PM
To: Tom Miller; Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] FTS 4060 Cesium Time & Frequency Standard, Datum
Update.
I received the FTS 4060 unit and guess what - it does not work.
So a few questions to the experts on this list-
1) How can I verify that the CBT is in fact bad? Is there any easy way to
confirm the high voltage supplies are working short of pulling the CBT
assembly out and making direct HV measurements? How about the other tube
functions? There is no beam activity on meter 3 and the ion pump voltage is
0.
2) If it is a bad tube (likely), is there any way it can be rebuilt? Does
anyone do such a thing?
3) Is there any source for used tubes?
4) If this is a dead end, is there any value in the rest of the unit? This
one has the display clock and the 5 MHz and 100 kHz options.
Or should I just bury it in the back yard and lick my wounds?
Regards,
Tom
----- Original Message -----
From: "Tom Miller" <tmiller at skylinenet.net>
To: "Discussion of precise time and frequency measurement"
<time-nuts at febo.com>
Sent: Tuesday, September 24, 2013 7:10 PM
Subject: [time-nuts] FTS 4060 Cesium Time & Frequency Standard, Datum
While we are on Cs standards, is there technical information (service
manual) available for the Datum FTS 4060 Cesium Time & Frequency Standard
and its options?
Thank you,
Tom
_______________________________________________
time-nuts mailing list -- time-nuts at febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
_______________________________________________
time-nuts mailing list -- time-nuts at febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
More information about the Time-nuts_lists.febo.com