[time-nuts] 10MHz LTE-Lite
Bob Camp
kb8tq at n1k.org
Sun Dec 7 14:38:30 UTC 2014
Hi
All GPSDO’s will have issues if you poke directly into them. The whole idea is to buffer the output(s) and keep them isolated from the outside world. That way the GPS can track out any errors and correct them. Any static error you see will be zeroed out if it has been there for long enough for the GPS to act on. Transients of any sort (load / voltage / temperature / gravity …) that are faster than the loop will show up on the output.
Bob
> On Dec 7, 2014, at 1:32 AM, Orin Eman <orin.eman at gmail.com> wrote:
>
> Hi Said,
>
> It's a little while since you sent this, but I just finished some testing
> with the LTE Lite.
>
> I already had a Trimble Thunderbolt and also have an HP 5335A with OCXO.
> The 5335A has shown the Trimble O/P 10 MHz +/- 0.03 Hz for the last few
> years (displayed frequency on the 5335A has drifted down by about 0.04 Hz).
>
> So, I set up the LTE Lite (10MHz TCXO version) with its antenna about 6'
> from that used by the Trimble. After letting it settle down, it looked
> good. I was using the high impedance input of the 5335A (aside: if set for
> DC coupling, the 5335A would read 20MHz from the ringing. I only have
> about 18" of RG-188 after the pigtail supplied with the LTE Lite).
>
> Today, I made up a buffer using a 74AC04 and LT1763-3.3 LDO regulator
> supplied with 5V from a bench supply. Two inverters each into 100 ohms
> then 0.1uF DC block as suggested. I connected this to the LTE Lite instead
> of directly to the 5335A with the 5335A set to 50 ohms and the 5335A read
> 0.5Hz low! It settled down to the original reading after a while. I
> looked at the signal from the buffer using a 50 ohm pass-through terminator
> and the signal looked nice and square.
>
> A few hours later I went back and it still looked good. I decided to look
> at the input to the buffer on a TDS-210 'scope with a 10X probe. There is
> now perhaps 6" of coax to the buffer which has 1 Mohm in parallel. The
> signal is about 5V pk-pk, but the ringing dies down quickly. BUT, the
> 5335A was displaying a few hundredths of a Hz higher than 10MHz, whereas it
> was a few hundredths below before! When I removed the probe, it went a few
> hundredths of a Hz below where it was originally and gradually recovered.
>
> So, in addition to temperature, the LTE-Lite eval board with 10MHz TCXO
> appears to be sensitive to load as well as temperature, such that just a
> 10X oscilloscope probe will affect the output. Normally, you probably
> wouldn't notice this, but I switched the loads while it was running.
> Certainly, once put in an enclosure, the 74AC04 buffer would be permanently
> connected and I'd assume any effect noticed above would be during the
> warmup of the LTE-Lite and wouldn't really be noticed.
>
> I'm sure I forgot some detail or other in the above description. I
> couldn't find a DIP 74AC04 so I made a PCB for an SMT 784AC04 using MG
> Chemicals 1/32" positive-resist PCB and Eagle for the design; 1206 100 ohm
> series resistors and 1206 0.1uF ceramic DC blocking capacitors. I laid it
> out for SMA sockets that you could wire RG-316 or similar directly instead.
> I used a 6" MMCX to RG-316 pigtail from the LTE-Lite to my buffer. I
> mounted one SMA socket on the outputs which is what I used to connect to
> the HP 5335A. The LT1763 has a 1uF 35V tantalum on the input and 10 uF 30V
> tantalum on the output (they are what I had in the parts bin, or I'd have
> used *smaller* ceramic parts). I tweaked the design for hand soldering
> with no solder mask (i.e. 20 mil clearances and top layer restrictions
> around most components to keep the ground fill away).
>
> Orin.
>
> On Sat, Nov 22, 2014 at 9:44 AM, Said Jackson via time-nuts <
> time-nuts at febo.com> wrote:
>
>> Hi Paul, Jim, David,
>>
>> Let me address all your emails:
>>
>> Glad you got your boards. $50 in overseas additional charges from your
>> post office sucks!
>>
>> Some hints for experimenting from what I have learned:
>>
>> You definitely want to build a 50 Ohms buffer for the 10MHz boards and the
>> synthesized outputs on all boards; on the 20MHz boards on the Tcxo output
>> it's optional.
>>
>> The biggest problem is building a suitable 3.3V or 5V power supply. I
>> built a buffer using two NC7SZ04 chips receiving the input in parallel with
>> a 1M terminating resistor to ground. Then using a 100 Ohms series resistor
>> on both outputs to get ~55 Ohms equivalent impedance, and combining the two
>> R's to drive the 50 Ohms inputs through a 100nF cap for DC blocking. You
>> can use a 74AC04 just as well. I tried a standard high quality 3-pin
>> regulator and got very bad AM noise modulation due to the large noise on
>> the rail. Then I used a very low noise LDO from LT and that solved the
>> problem and the output is now very clean and drives 50 Ohms inputs with
>> ease.
>>
>> you can grab the very low noise 3.0V rail output from the eval board to
>> power the buffers (see the schematics in the user manual) but loading that
>> creates a bit of heat on the LTE-Lite which will affect stability a tiny
>> bit.
>>
>> On the power consumption, you can see we go through a linear regulator to
>> get 3.3V from 5V USB/EXT power. This is very inefficient. For best power
>> consumption you want to use a very high efficiency buck regulator to
>> generate the 3.3V from your battery. This means you loose the USB interface
>> though as that chip runs from 5V and has an internal LDO.
>>
>> On the zero Ohms R2/R3 resistors, check the schematics - these allow you
>> to power the DIP-14 Tcxo from either the digital 3.3V rail, or the
>> low-noise 3.0V rail (default). The software will auto-detect if you attach
>> a 10MHz or 20MHz Tcxo, no configuration is needed.
>>
>> On drafts, yes that is the number one cause of phase drifts. We put the
>> board into an ESD bag, and put some thin ESD padding material on top. That
>> prevents drafts, and following the EFC curve you can see the unit still
>> reacts slowly to the AC or heaters kicking on. That's normal, and that's
>> why we discipline to GPS.. In our setup the units have typically less than
>> 20ns standard deviation on the RF and 1PPS outputs. The raw gps 1PPS output
>> on the header is even better on average, but has the sawooth error on it.
>> The sawtooth error correction value is in the PSTI NMEA message for those
>> that want to use the raw gps 1PPS output and correct the sawtooth
>> externally. This chip has a very high rate internal system frequency that
>> results in very low residual sawtooth error.
>>
>> On the auto survey process - this is disabled when using 3D mobile mode by
>> shorting pins 1 and 3 on the 3-pin header as described in the "read me
>> first". But be aware that changing that header with power applied results
>> in flash memory corruption, and thus a very bad day. That's why we did not
>> solder the header - to avoid any accidents..
>>
>> Bye,
>> Said
>>
>>
>>
>>
>>
>> Sent from my iPad
>>
>> On Nov 22, 2014, at 8:42, paul swed <paulswedb at gmail.com> wrote:
>>
>>> Ok give it a week for the magic to wear off.Then its time to hack.
>>> I am sort of headed into that mode.
>>> The system draws what I would technically call squat for power. Hmm
>> wonder
>>> how thats measured VA watts??
>>> Locks pretty darn fast and recovers pretty fast. But you do always go
>>> through the survey. Not really a negative.
>>> It does produce a relatively stable output. But you can see it slip as
>>> compared to the likes of the Z3801 or 3811. This tends to be due to
>> drafts.
>>> Its quite sensitive.
>>> So following best practices two thick socks are on top of it now.
>>>
>>> So to the hacking/curiosity.
>>> I will group my interests in several areas (No particular sequence);
>>>
>>> - A good enough reference to replace my cheap-y $25 Telco RB reference.
>>> - Adding the various buffers to have useful signals.
>>> - Trying to keep the support power consumption down to match the LTE.
>>> - The curiosity of adding a oven 10 Mhz oscillator. I have a PTI and
>>> 10811
>>> - Battery back up a real question given the lockup time of the unit.
>>>
>>>
>>> The buffering and dividing will come first and may have to be 74 HC or
>> HCT
>>> to get the 20 to 10 Mhz. Its unfortunate because 74 AC would allow
>>> everything to run on 3.3V. May just wait and order some AC chips and do
>> it
>>> right.
>>>
>>> Then LPF the output 10 Mhz to a sine wave and hit a transistor buffer.
>>> The buffer will be the biggest power pig of all. They always are.
>>>
>>> I always seem to need various ticks. The 1 PPS will be adapted to RS232
>> and
>>> RS 485 using buffers/converters. Simple 1 chip wonders.
>>> The output only data feed could also be RS232 and there is a spare
>>> transmitter in the max chip I would use. These can be wired on a board or
>>> for almost nothing ordered from ebay these days with shipping delays.
>>>
>>> The oven is really a curiosity. I have a 20 MHz unit. I think that by
>>> changing the 3 zero ohm Rs on the system I can shift to 10 Mhz and
>> directly
>>> replace the 20 Mhz TCO. If thats not true then the typical oven has to be
>>> doubled to 20 and then converted to a clean 3V digital signal. This
>> thread
>>> already has some hints on the EFC voltage.
>>>
>>> Lastly Battery backup. By the time I get to here I will have decided if
>> its
>>> even worth the effort. Batteries are a pain in the .... But is nice in
>> that
>>> the system just runs. Having an oven absolutely takes a 1-2 watt solution
>>> to a 30 watt total solution. Essentially what my RB consumes today when
>> you
>>> look at the wasted energy in the transformers and such. I am using an HP
>>> battery system that drove RB and CS references circa 1980. So not very
>>> efficient. But sure does work.
>>>
>>> Someplace sooner then later a box for it all. Drafts do upset the TCXO.
>> It
>>> may need to be a highly customized temporary box. These boxes are
>> available
>>> at most supermarkets. Ask for cardboard.
>>>
>>> So there you have it my 10 cents worth of musings on the direction I am
>>> headed.
>>> Reagards
>>> Paul
>>> WB8TSL
>>>
>>>
>>>
>>>
>>>
>>>
>>>
>>>
>>>
>>>
>>>
>>> On Sat, Nov 22, 2014 at 10:37 AM, Jim Sanford <wb4gcs at wb4gcs.org> wrote:
>>>
>>>> Said:
>>>> Just ordered a second 10 MHz board for my rover station....
>>>> 73,
>>>> Jim
>>>> wb4gcs at amsat.org
>>>>
>>>>
>>>> On 11/20/2014 3:32 PM, S. Jackson via time-nuts wrote:
>>>>
>>>>> Hello everyone,
>>>>> after what must have been the longest thread in T-nuts history its
>>>>> almost
>>>>> all quiet today. I am going to take advantage of that and announce
>> some
>>>>> good news:
>>>>> Its a miracle: the 10MHz DIP-14 TCXOs for the LTE-Lite came in weeks
>>>>> ahead
>>>>> of schedule from the factory! And they work very well.
>>>>> We will thus be shipping out the 10MHz LTE-Lite eval boards in the
>> next
>>>>> couple of working days. There are still a number left for sale on Ebay
>>>>> (search for "LTE Lite GPSDO"), so if you were hesitant to get one due
>> to
>>>>> the long
>>>>> lead-time, then now is your chance.
>>>>> Also, after being in time nuts hands for almost a week I am surprised
>>>>> there are very few mails, questions, or comments about the 20MHz
>> boards,
>>>>> and we
>>>>> have received almost no feedback on Ebay :( I hope that is a good
>> sign.
>>>>> Bye,
>>>>> Said
>>>>>
>>>>>
>>>>>
>>>>> _______________________________________________
>>>>> time-nuts mailing list -- time-nuts at febo.com
>>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/
>>>>> mailman/listinfo/time-nuts
>>>>> and follow the instructions there.
>>>>>
>>>>>
>>>>
>>>> ---
>>>> This email has been checked for viruses by Avast antivirus software.
>>>> http://www.avast.com
>>>>
>>>> _______________________________________________
>>>> time-nuts mailing list -- time-nuts at febo.com
>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/
>>>> mailman/listinfo/time-nuts
>>>> and follow the instructions there.
>>>>
>>> _______________________________________________
>>> time-nuts mailing list -- time-nuts at febo.com
>>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>> and follow the instructions there.
>> _______________________________________________
>> time-nuts mailing list -- time-nuts at febo.com
>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>>
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
More information about the Time-nuts_lists.febo.com
mailing list