[time-nuts] Did a member of time-nuts buy this?
Bob Camp
kb8tq at n1k.org
Tue Dec 16 01:30:34 UTC 2014
Hi
Here’s another way to sort this all out:
For ADEV, you probably want to get down to ~ 1x10^-13 at 1 second. If the system improves by 1/tau, it will be adequate for anything else you want to do. It’s a system that would need to resolve 100 fs, so most of the “counter” devices that die out around 10 ps would be ruled out.
For long term, you might decide on 1x10^-15 at 100,000 seconds. If it also goes as 1/tau you would have 1x10^-13 at 1,000 seconds. Not good enough for ADEV, but quite adequate for things like aging. Any of the 10 to 100 ps range devices would do fine in this case.
Sort into two piles and then sub-sort.
Bob
> On Dec 15, 2014, at 2:27 AM, Mike Monett <timenuts at binsamp.e4ward.com> wrote:
>
>> Hi
>
>>> On Dec 10, 2014, at 11:12 PM, Mike Monett
>>> <timenuts at binsamp.e4ward.com> wrote:
>
> [...]
>
>>>>> Can you tell me some of the ones that do?
>
>>>> I have yet to see one for under $2K that does it correctly. I
>>>> don't have the cash to buy ones at those sort of prices. Some
>>>> have reported that the old Motorola UT's will do it. The samples
>>>> I have tried have not done very well. I may have not had them
>>>> running right who knows.
>
>>> OK, that pretty much eliminates the NIST data.
>
>> or means that more research is needed on single sat boards. There
>> are a *lot* of them out there.
>
> What is a "single sat board" and can you give some examples?
>
>>>>>> 5) Feed that into your control loop equation.
>
>>>>> There's another term I need to research!
>
>>>> It's not a simple control process, but it's not all that
>>>> terrible either. It just takes a bit of work to optimize. Figure
>>>> a few months to a few years for the optimization depending on
>>>> what you have for issues along the way.
>
>>>>> OK, so I figure out how to do this. How do I tell if this is
>>>>> making the gpsdo more accurate? In other words, how do I get
>>>>> the ADEV without having an H-Maser?
>
>>>> You get a Cs (or other atomic standard) or you compare several
>>>> different GPSDO's against each other. The preference would be
>>>> for groups of three so you can rule out ones that are not doing
>>>> what they should.
>
>> The ideal would be > 3 TBolts, > 3 same model Rb's, > 3 same model
>> OCXO's and more than three groups overall. Each has their own ADEV
>> curve. Comparing devices with vastly different ADEV is not the
>> best way to go.
>
> Yes, that makes good sense.
>
> [...]
>
>>> There are many new dacs, op amps, and bipolar microwave devices
>>> that offer much lower noise than current designs use. I think the
>>> performance can be improved in some areas with new components and
>>> design techniques, and I have equipment and time to explore.
>
>> There are only two points that system noise really comes into the
>> GPSDO design:
>
>> 1) The TDC must have enough resolution
>
>> 2) The DAC on the EFC must have noise below the OCXO
>
> I was more thinking of the Rubidium physics package. Most of the
> Rubidiums on eBay are pretty old.
>
> The lamp must have constant intensity. The null detector has to
> identify a very shallow drop. The microwave signal needs to have
> excellent phase noise.
>
> These all need low noise, low flicker components, and significant
> advances have been made in recent years. The surrounding circuitry
> could probably benefit from a redesign to take better advantage of
> low noise components. For example, SRD's are much noisier than NLTL's.
>
> Probably little can be done to improve the Rubidium cell, but it
> should be the limiting factor and not the electronics.
>
>> The TDC is limited by the basic resolution of the GPS system. It's
>> easy to build one that has far more resolution than needed /
>> useful.
>
> Some people have little faith:)
>
>> The DAC issue is normally solved with a < $4 part. Unless you have
>> an OCXO with a very wide EFC range, it's noise is unlikely to be
>> an issue. DAC resolution can be addressed to any desired level
>> with two DAC's (fine and coarse). (Once you get going, you only
>> use the fine DAC).
>
>> The real "fun and games" revolves around the software used to
>> implement the filtering / control loop between the GPS and the
>> OCXO (or Rb, or TCXO, or MEMS, or VCXO, or Cs or âŠ)
>
>> Focus on the software.
>
>> Bob
>
> I have a few Rubidiums and OCXOs I'd like to get running for a month
> or so to stabilize. During this time, I'd like to monitor the
> performance to discover any bad units and see which are the best
> ones.
>
> For example, TVB shows a 100:1 variation in ADEV in FE-405B
> Rubidiums. The section is titled "Variation in FE-405B" in
>
> http://leapsecond.com/pages/fe405/
>
> Clearly, it would be futile to try to use a bad unit in setting up a
> gpsdo.
>
> I need some means of measuring the performance of these units while
> they are running. A dozen HP5370's would be out of the question. I
> did some research to find the different methods available and decide
> which has the lowest per-channel cost and best performance.
>
> Here are some of the references I found. I discarded most of the poor ones
> and tried to keep only the ones that talk about measurements in the
> picosecond or femtosecond range.
>
> I did not include DMTD since the concept is so simple. The Reviews help to
> get oriented, but sometimes it takes reading the papers and the patents to
> see the timing diagrams and understand what the author is trying to do.
>
> There are many different variations on the FPGA approach. I would be
> concerned about the development time, the large DNL, and the problems with
> crosstalk on multi-channel units.
>
> The TI THS788 looks good on paper, but it is single-source, not well
> stocked, and there is lttle information on crosstalk between channels. It
> is also quite expensive per channel. There are no application notes and
> little or no information on usage on the web.
>
> The Thesis generally have excellent reviews worth reading.
>
> Articles
>
> Simple PICTIC 250ps time interval counter
> http://www.ko4bb.com/dokuwiki/doku.php?id=precision_timing:pictic
>
> Griffiths Time to Digital converters
> http://www.ko4bb.com/~bruce/TDC.html
>
> [time-nuts] Experience with THS788 from TI?-
> Attila Kinali's post on TDC methods
> https://www.febo.com/pipermail/time-nuts/2012-March/065337.html
>
> Reviews
>
> Review of methods for time interval measurements with picosecond
> resolution
> http://ztc.wel.wat.edu.pl/met4_1_004.pdf
>
> Low resource FPGA-based 32 Channel Time to Digital Converter
> - Excellent review of TDC techniques and measurement verification
> http://arxiv.org/vc/arxiv/papers/1206/1206.0679v3.pdf
>
> Design, construction and tests of a high resolution, high dynamic
> range Time to Digital Converter
> - Excellent review of TDC methods and test techniques
> http://inspirehep.net/record/1313667/files/getfile.pdf
>
> Review Of Sub-Nanosecond Time-Interval Measurements
> http://www.slac.stanford.edu/cgi-wrap/getdoc/slac-pub-1331.pdf
>
> Chapter 1 Time Interval Measurement Literature Review
> http://www.rrsg.ee.uct.ac.za/members/jon/activities/timcs.pdf
>
> Time intervals measurements and generation methods review
> http://www.ohwr.org/attachments/128/Time_Interval_Measurements_Techniques.pdf
>
> Papers
>
> A Cyclic CMOS Time-to-Digital Converter With Deep Sub-nanosecond
> Resolution
> http://ntur.lib.ntu.edu.tw/bitstream/246246/2007041910021548/1/00777354.pdf
>
> A 96-Channel FPGA-based Time-to-Digital Converter
> http://arxiv.org/pdf/physics/0502062.pdf
>
> Complete and Compact 32-Channel System for Time-Correlated
> Single-Photon Counting Measurements
> http://ieeexplore.ieee.org/ielx7/4563994/6588410/06617673.pdf?arnumber=6617673
>
> Design and Development of GPS Receiver for PNSS-1
> http://www.suparco.gov.pk/pages/presentations-pdf/day-1/session-2/12C-II/1.pdf
>
> A Novel Ultra-fast High Resolution Time-domain EMI Measurement
> System based on Field Programmable Gate Arrays
> https://mediatum.ub.tum.de/doc/1163324/1163324.pdf
>
> High Precision Frequency Measurement System Based on Different
> Frequency Phase Processing (in Chinese)
> http://xbzrb.tjujournals.com/oa/pdfdow.aspx?Type=pdf&FileName=11acd53a-876d-4ab0-8906-572aa3836215.pdf
>
> A 51-dB SNDR DCO-Based TDC Using Two-Stage Second-Order Noise Shaping
> http://www28.cs.kobe-u.ac.jp/pdf/1205_konishi_iscas.pdf
>
> Integrated High-Resolution Multi-Channel Time-to-Digital Converters
> (TDCs) for PET Imaging
> - Discusses 100 fs Gated-ring-oscillator (GRO)
> http://cdn.intechopen.com/pdfs-wm/12913.pdf
>
> A 26 ps RMS time-to-digital converter core for Spartan-6 FPGAs
> http://arxiv.org/pdf/1303.6840v1.pdf
>
> A High-Resolution Flash Time-to-Digital Converter Taking Into
> Account Process Variability
> http://conferences.computer.org/async2007/PRS/15-minas-async07.pdf
>
> Principles and Instrumentation in Time-of-flight Mass Spectrometry
> http://www.wiley.com/legacy/wileychi/ms/articles/1519_a.pdf
>
> TEMPERATURE SENSING CRYSTAL HTS-206
> - maybe more stable than thermistor
> http://www.abcelectronique.fr/composants/telechargement_datasheet.php?id=385911&part-number=HTS-206
>
> A novel method of measurement of time and amplitude of analog
> signals based solely on FPGA units
> http://koza.if.uj.edu.pl/pet-symposium-2013/talks/17.pdf
>
> A 16 channel high resolution (<11 ps RMS) Time-to-Digital Converter
> in a Field Programmable Gate Array
> http://iopscience.iop.org/1748-0221/7/02/C02004/pdf/1748-0221_7_02_C02004.pdf
>
> A 128-channel Time-to-Digital Converter (TDC) inside a Virtex-5 FPGA
> on the GANDALF module
> http://iopscience.iop.org/1748-0221/7/03/C03008/pdf/1748-0221_7_03_C03008.pdf
>
> FPGA Based Data Digitisation with Commercial Elements
> http://indico.scc.kit.edu/indico/getFile.py/access?contribId=4&sessionId=4&resId=0&materialId=slides&confId=21
>
> Time-to-Digital Converter Architecture with Residue Arithmetic and
> its FPGA Implementation
> http://www.el.gunma-u.ac.jp/~kobaweb/news/pdf/2014/2014-11li_congbing2.pdf
>
> An FPGA-Integrated Time-to-Digital Converter Based on a Ring
> Oscillator for Programmable Delay Line Resolution Measurement
> - 0.4 ps Xilinx FPGA
> http://downloads.hindawi.com/journals/jece/2014/230803.pdf
>
> First Order Noise Shaping Local-Oscillator Based Time-to-Digital Converter
> - Gated Ring Oscillator based TDC
> http://ims.unipv.it/~franco/ConferenceProc/313.pdf
>
> A High Resolution FPGA-based TDC with Nonlinearity Calibration
> http://www2.ess.nthu.edu.tw/~cwlu/conference/5_A%20High%20Resolution%20FPGA-based%20TDC%20with.pdf
>
> A Single-Slope 80MS/s ADC using Two-Step Time-to-Digital Conversion
> http://dspace.mit.edu/openaccess-disseminate/1721.1/60073
>
> Simulation of ps-Detector Electronics
> http://hep.uchicago.edu/workshops/2005-picosecond/8-Fukun_Tang_ps_sim.ppt
>
> High Precision Time and Frequency Counter for Mobile Applications
> http://www.wseas.us/e-library/transactions/circuits/2010/89-747.pdf
>
> Time-to-Digital Converter based on Analog Time Expansion for 3D Time
> of Flight Cameras
> https://pure.ltu.se/portal/files/95016947/90220A.pdf
>
> A Pc-Based Time Interval Counter With 200 Ps Resolution
> http://www.dtic.mil/get-tr-doc/pdf?AD=ADA427911
>
> The Development of Large-area, Picosecond Resolution, Time-of-Flight
> Detectors
> http://hep.uchicago.edu/~frisch/adr.pdf
>
> Low-Voltage High-Frequency Gated Ring Oscillator Using Bootstrap
> Technique
> http://www.el.gunma-u.ac.jp/~kobaweb/news/pdf/2012/AMDE2012_Kamiyama.pdf
>
> High-resolution short time interval measurement system implemented
> in a single FPGA chip
> http://download.springer.com/static/pdf/727/art%253A10.1007%252Fs11434-011-4421-3.pdf?auth66=1418360984_e35e8a8588d3977c4d4d4d2109c1aaaa&ext=.pdf
>
> Implementation of sub-nanosecond time-to-digital convertor in
> field-programmable gate array
> https://hal.archives-ouvertes.fr/in2p3-00873192/PDF/articleMST_Marteau-et-al_v1.pdf
>
> New Jitter Measurement Technique Using TDC Principle in a FPGA
> Component
> http://www.inatel.br/biblioteca/component/docman/doc_download/3831-new-jitter-measurement-technique-using-tdc-principle
>
> High Resolution Modular Time Interval Counter
> http://ilrs.gsfc.nasa.gov/docs/timing/artyukh_time_interval_counter.pdf
>
> Using Dynamic Reconfiguration to Implement High-Resolution
> Programmable Delays on an FPGA
> http://www.iro.umontreal.ca/~feeley/papers/BergeronFeeleyDaigneaultDavidNEWCAS08.pdf
>
> A 2.6ps rms -Period-Jitter 900MHz All-Digital Fractional-N PLL Built
> with Standard Cells
> http://www-bsac.eecs.berkeley.edu/publications/search/send_publication_pdf2client.php?pubID=1307595132
>
> C111 / P111 4 channel Time to Digital Converter User's manual
> http://www.esrf.eu/files/live/sites/www/files/Instrumentation/DetectorsAndElectronics/Units/Electronics/digital-electronics-lab/c111.pdf
>
> Carry-chain propagation delay impacts on resolution of FPGA-based TDC
> http://www.j.sinap.ac.cn/nst/EN/article/downloadArticleFile.do?attachType=PDF&id=457
>
> A 19.6 ps, FPGA-Based TDC With Multiple Channels for Open Source
> Applications
> http://ens.ewi.tudelft.nl/pubs/charbon12tcas_favi.pdf
>
> A data driven high performance Time to Digital Converter
> http://ph-collectif-lecc-workshops.web.cern.ch/ph-collectif-lecc-workshops/LEB00_Book/tracker/christiansen_track.pdf
>
> On-Chip Timing Measurement Architecture with Femtosecond Resolution
> http://eprints.soton.ac.uk/262543/1/collinsm_timemeasurement2.pdf
>
> A 13b 315fs rms 2mW 500MS/s 1MHz Bandwidth Highly Digital
> Time-to-Digital Converter Using Switched Ring Oscillators
> http://www.cse.psu.edu/~xydong/files/proceedings/ISSCC2012/files/D27_04.pdf
>
> A high resolution Time-to-Digital Converter on FPGA for
> Time-Correlated Single Photon Counting
> http://libgen.org/scimag/get.php?doi=10.1109/mwscas.2012.6292166
>
> A Design of Vernier Coarse-Fine Time-to-Digital Converter using
> Single Time Amplifier
> http://ocean.kisti.re.kr/downfile/volume/ieek/E1STAN/2012/v12n4/E1STAN_2012_v12n4_411.pdf
>
> All Digital Time-to-Digital Converter with High Resolution and Wide
> Detect Range
> http://www.engineeringletters.com/issues_v19/issue_3/EL_19_3_16.pdf
>
> Examples of 1 PPS Clock Measuring Systems
> http://www.wriley.com/Examples%20of%201%20PPS%20Clock%20Measuring%20Systems.pdf
>
> On-Chip Processing for the Wave Union TDC Implemented in FPGA
> http://lss.fnal.gov/archive/2009/conf/fermilab-conf-09-275-e.pdf
>
> 18-channel FPGA Wave Union TDC for Time-of-Flight Applications
> http://lss.fnal.gov/archive/2009/conf/fermilab-conf-09-573-ppd.pdf
>
> A 96-channel FPGA-based Time-to-Digital Converter (TDC)
> http://www.phys.sinica.edu.tw/~e906/WWW_public/Doc/TDC/FPGA_TDC.pdf
>
> Design of FPGA-based TDC for the ICAL Detector of India-based
> Neutrino Observatory
> http://www.hecr.tifr.res.in/~sudeshnadg/publication/FPGA_TDC.pdf
>
> FPGA based Time-to-Digital Converter
> http://www.sympnp.org/proceedings/56/G7.pdf
>
> A Flash Time-to-Digital Converter with Two Independent Time Coding
> Lines
> IMEKO-IWADC-2011-12.pdf
> http://www.imeko.org/publications/iwadc-2011/IMEKO-IWADC-2011-12.pdf
>
> A 17ps Time-to-Digital Converter Implemented in 65nm FPGA Technology
> http://infoscience.epfl.ch/record/139431/files/isfpga09-cfavi.pdf
>
> A 12-Bit Vernier Ring Time-to-Digital Converter in 0.13 um CMOS Technology
> http://www.eng.auburn.edu/~daifa01/Top/PubPapers/2010/Jpaper2010-02.pdf
>
> A high-speed wide dynamic range time-to-digital converter
> http://www.ssl.berkeley.edu/~mlampton/LamptonRaffanti.pdf
>
> High Resolution Time-Interval Measurement Systems Applied To Flow
> Measurement
> http://www.metrology.pg.gda.pl/full/2014/M&MS_2014_077.pdf
>
> Accurate time interval measurement electronics for pulsed time of
> flight laser radar
> http://www.ee.oulu.fi/~kari/Papers/Nantes1997.pdf
>
> Signal Processing for Pico-second Resolution Timing Measurements
> http://psec.uchicago.edu/Papers/NIM_v8b_final.pdf
>
> High-Resolution Time-to-Digital Converter in Field Programmable Gate Array
> http://cds.cern.ch/record/1158663/files/p383.pdf
>
> Random Errors in Time Interval Measurement Based on SAW Filter Excitation
> http://libgen.org/scimag/get.php?doi=10.1109/tim.2007.915465
>
> A 96-channel, 500 ps resolution TDC board for the BaBar experiment at SLAC
> http://www.ge.infn.it/babar/ITB/RT99/paper.pdf
>
> A Cyclic Vernier TDC for ADPLLs Synthesized From a Standard Cell Library
> http://wwweb.eecs.umich.edu/wics/publications/Park_TCASI2011.pdf
>
> Nonlinearity Correction of the Integrated Time-to-Digital Converter
> with Direct Coding
> http://libgen.org/scimag/get.php?doi=10.1109/19.571882
>
> PICTIC Interpolator Linearity
> http://www.wriley.com/PICTIC%20Interpolator%20Linearity.pdf
>
> Firmware-only Implementation of Time-to-Digital Converter (TDC) in
> Field-Programmable Gate Array (FPGA)
> http://www-ppd.fnal.gov/EEDOffice-W/Projects/ckm/comadc/PID26522.pdf
>
> High-Precise Portable Time Interval and Frequency Counter
> http://tycho.usno.navy.mil/ptti/2012papers/PTTI2012.p247.pdf
>
> FPGA-based time counters with a wander measurement mode
> http://www.pwt.et.put.poznan.pl/PWT_2012/PWT%202012_2242.pdf
>
> The Simple PICTIC
> http://www.ko4bb.com/dokuwiki/lib/exe/fetch.php?id=precision_timing%3Apictic&cache=cache&media=precision_timing:simple_pictic.pdf
>
> A Saw-tooth Wave Based Design of Time to Digital Converter
> http://www.wseas.us/e-library/conferences/2009/budapest/SMO/SMO68.pdf
>
> A Multi-Path Gated Ring Oscillator TDC With First-Order Noise Shaping
> http://www.cppsim.org/Publications/JNL/straayer_jssc09.pdf
>
> Micrel LVPECL, CML 5GHz - 7GHz Buffer
> http://www.micrel.com/_PDF/Eval-Board/sy58011-12-13u_eb.pdf
>
> Micrel LVPECL, CML 4GbpS - 10.7Gbps Multiplexer
> http://www.micrel.com/_PDF/Eval-Board/sy58017-18-19u_eb.pdf
>
> A flash high-precision Time-to-Digital Converter implemented in FPGA
> technology
> http://indico.cern.ch/event/49682/session/35/contribution/3/material/slides/0.ppt
>
> FPGA-Based High Area Efficient Time-To-Digital IP Design
> http://www.ifm.umich.mx/~villasen/TESIS/Mario/TDC.pdf
>
> Firmware-only Implementation of Time-to-Digital Converter (TDC) in FPGA
> http://www-ppd.fnal.gov/EEDOffice-W/Projects/ckm/comadc/tdc1s.pdf
>
> A Time-To-Digital Converter (TDC)
> https://indico.phys.hawaii.edu/getFile.py/access?contribId=4&resId=0&materialId=slides&confId=392
>
> 96-Channel FPGA-Based Time-To-Digital Converter
> http://hep.uchicago.edu/~frisch/TDC_NIM_v12.pdf
>
> Texas Instruments THS788 Quad-Channel Time Measurement Unit (TMU)
> http://www.ti.com/lit/ds/symlink/ths788.pdf
>
> A 128-channel Time-to-Digital Converter (TDC) inside a Virtex-5 FPGA
> on the GANDALF module
> http://wwwhad.physik.uni-freiburg.de/gandalf/modules/download_gallery/dl.php?file=59
>
> Vernier's Delay Line Time-to-Digital Converter
> http://www.np.ac.rs/yu/preuzimanjasve/publications/vol1br1/181-verniers-delay-line-time-to-digital-converter/download
>
> Time-to-Digit Converter Based on Radiation-Tolerant FPGA
> - Interesting calibration analysis
> http://kmlinux.fjfi.cvut.cz/~vacekmic/docs/vzlu_tdc.pdf
>
> A High-Resolution Time-to-Digital Converter Implemented in
> Field-Programmable-Gate-Arrays
> http://mi.sibet.cas.cn/zlgx/201310/W020131024537685817071.pdf
>
> Picosecond Resolution Time-to-Digital Converter Using G m -C
> Integrator and SAR-ADC
> http://www.ssc.pe.titech.ac.jp/publications/2014/IEEE_trans/xu/Xu_GmTDC_Journal_Final_V5.pdf
>
> A Design of Vernier Coarse-Fine Time-to-Digital Converter using
> Single Time Amplifier
> http://www.jsts.org/html/journal/journal_files/2012/12/Year2012Volume12_04_05.pdf
>
> Patents
>
> Burbeck METHOD AND APPARATUS FOR AUTOMATICALLY MEASURING TIME
> INTERVALS
> https://patentimages.storage.googleapis.com/pdfs/US2665410.pdf
>
> Frady DOUBLE INTERPOLATION METHOD AND APPARATUS FOR MEASURING TIME
> INTERVALS
> https://patentimages.storage.googleapis.com/pdfs/US2665411.pdf
>
> Bagley ELECTRONIC INTERPOLATING COUNTER FQR THE TIME INTERVAL AND
> FREQUENCY MEASUREMENT
> https://patentimages.storage.googleapis.com/pdfs/US3133189.pdf
>
> Nutt DIGITAL INTERVALOMETER
> https://patentimages.storage.googleapis.com/pdfs/US3983481.pdf
>
> Janowitz TIME INTERVAL MEASUREMENT METHOD AND APPARATUS
> https://patentimages.storage.googleapis.com/pdfs/US3999128.pdf
>
> Thornton HIGH RESOLUTION DATA ACQUISITION
> http://patentimages.storage.googleapis.com/pdfs/US5200933.pdf
>
> Condreva HIGH RESOLUTION TIME INTERVAL COUNTER
> https://patentimages.storage.googleapis.com/pdfs/US5333162.pdf
>
> Kattan TIME INTERVAL ANALYZER HAVING MULTIPLE MEASUREMENT CIRCUITS
> http://patentimages.storage.googleapis.com/pdfs/US6226231.pdf
>
> Nair CORRECTION FOR PIPELINED ANALOG TO DIGITAL (A/D) CONVERTER
> http://patentimages.storage.googleapis.com/pdfs/US6784814.pdf
>
> Gao HIGH RESOLUTION SAMPLING-BASED TIME TO DIGITAL CONVERTER
> http://patentimages.storage.googleapis.com/pdfs/US8564471.pdf
>
> Thesis
>
> An Integrated Cmos High Precision Time-To-Digital Converter Based On
> Stabilised Three-Stage Delay Line Interpolation
> - Excellent review
> http://herkules.oulu.fi/isbn951427461X/isbn951427461X.pdf
>
> Multi-Path Differential Delay Line based Time-to-Digital Converter
> for ADPLL
> http://www.diva-portal.org/smash/get/diva2:635168/FULLTEXT01.pdf
>
> Noise Shaping Techniques for Analog and Time to Digital Converters
> Using Voltage Controlled Oscillators
> http://cppsim.org/Publications/Theses/straayer_phdthesis.pdf
>
> Implementation of the high resolution high dynamic Time to Digital
> Converter
> https://www.politesi.polimi.it/bitstream/10589/92658/1/Thesis_V1.0.2_Build04031059.pdf
>
> A Stabilized Multi- Channel Cmos Time-To- Digital Converter Based On
> A Low Frequency Reference
> http://herkules.oulu.fi/isbn9789514299322/isbn9789514299322.pdf
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
More information about the Time-nuts_lists.febo.com
mailing list