[time-nuts] HP 53131A - reset security code
Tom Van Baak
tvb at LeapSecond.com
Wed Dec 24 01:34:51 UTC 2014
Ulli,
You're right. The actual procedure is not described in the manual(s):
Operating Guide
http://cp.literature.agilent.com/litweb/pdf/53131-90055.pdf
Programming Guide
http://cp.literature.agilent.com/litweb/pdf/53131-90044.pdf
Assembly-Level Service Guide
http://cp.literature.agilent.com/litweb/pdf/53131-90039.pdf
Component Level Information
http://literature.cdn.keysight.com/litweb/pdf/5989-6308EN.pdf
But I have three ides for you.
1) The last PDF shows U14 is AT28C64B (www.atmel.com/Images/doc0001h.pdf) so if all else fails, you can use a logic analyzer on the address(13) / data(8) pins and see what's in the EEPROM. With some experimentation by you or others on the list we can locate the calibration security. This has the advantage of getting the existing code, without having to clear the entire EEPROM. Some other HP bench instruments use an onboard jumper to override the cal lock, but I don't see that on a 53131A. So perhaps there is a magic front panel sequence or SCPI command to clear NVRAM.
2) Search for 53132 at http://www.ko4bb.com/manuals/index.php and grab the 4 EPROM's. Do a 16-bit merge of U8/U10 and U9/U11 and look for ascii strings. All the SCPI words are in there and not all are documented. For example near the words SYST and CODE and SEC[URITY] is ALL and DEL[ETE]. Probably safe to try. Lots of interesting output strings, both for GPIB and for internal debugging. Even the words EEPROM CLEAR, along with "Oooh! That tickles. I don't know that vector." which we also see in the Z3801A, etc.
There's also a hidden "SERVICE MENU", complete with Easter egg:
MLB TEAM KRISTI BITTNER, LEE COSART, ERIC INGMAN, GARY JACOBSEN, SOOK-LAN LOH, ROB MITCHELL, ROBIN POSKUS, DON SCHREMP, DANA STOFFERS, AND GEORGE TSAI
SOFTWARE L. COSART G. JACOBSEN S. LOH R. POSKUS D. STOFFERS
HARDWARE K. BITTNER E. INGMAN D. SCHREMP
MECHANICAL G. TSAI
PROJECT MGR R. MITCHELL
BASEHIT -- 53132
3) There's a pForth interpreter in this instrument too. Note pr_eeprom_debug and wr_eeprom.
That's enough hints to keep someone on the list (Magnus?) busy over Christmas. Of course, if anyone knows the "confidential procedure", they should just let us know...
Thanks,
/tvb
----- Original Message -----
From: "Mod Mix" <modmix at t-online.de>
To: <time-nuts at febo.com>
Sent: Tuesday, December 23, 2014 3:04 PM
Subject: [time-nuts] HP 53131A - reset security code
> Hi,
>
> I got a used HP 53131A to which I added a stability option board (ocxo).
> In order to calibrate the device, a security code is needed - I don't
> know it. The seller doesn't know it. It's not the facorty set code.
> Advice on how the reset the non-volatile memory would be much appreciated.
>
> Thank you very much & happy Christmas
>
> Ulli
>
> From the 53131A service guide:
> A calibration security code feature allows you to enter a security code
> (electronic key) to prevent accidental or unauthorized calibrations of
> the Counter. The security code is set to 53131 or 53132 (corresponding
> to the model) when the Counter is shipped from the factory.
> If you forget your security code, you can reset the security code to the
> model-number default by resetting all of the non-volatile memory to a
> default state (Contact Agilent Technologies for this confidential
> procedure).
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
More information about the Time-nuts_lists.febo.com