[time-nuts] schematics of frequency counter
Li Ang
lllaaa at gmail.com
Mon Dec 29 12:55:53 UTC 2014
Hi Magnus,
The unit of these data is not ns but reference clock cycles (100ns).
TDC_GP22 measures the time between the edge of tdc_start and
tdc_stop1, then it measures the reference clock automaticly. The result you
get from it is the ratio of them.
2014-12-29 19:58 GMT+08:00 Magnus Danielson <magnus at rubidium.dyndns.org>:
> Hi,
>
> Some quick statistic-processing.
>
> Histogram of your data:
> 1.979 0
> 1.980 2
> 1.981 46
> 1.982 173
> 1.983 523
> 1.984 1031
> 1.985 1301
> 1.986 1131
> 1.987 648
> 1.988 236
> 1.989 8
> 1.990 1
> 1.991 0
>
> The total sample count is 5100 (wc -l only gives 5099 since there is a
> missing end of line, but word count is 5100).
>
> The average is about 1,985075 ns and it is reasonably gaussian, but with
> some systematics, notice how the slope is more abrupt on the higher end
> than the lower end. A quick and dirty spreadsheet gives me about 2,243 ps
> RMS jitter, which isn't all that bad. Yes, it will spread out to that 11 ps
> peak-to-peak jitter, but it is to be expected.
>
> It's pretty respectable for a home-built.
>
> Cheers,
> Magnus
>
>
> On 12/28/2014 03:19 PM, Li Ang wrote:
>
>> Hi Bob,
>> I did some test according to your suggestions. DUT is a symmetricom
>> x72
>> rb oscillator. Also, I've tried signal generator as the DUT. R&S SMY01 is
>> not as good as HP8662A but that the best I've got. The signal geneator is
>> also using FE5650 as ref clock.
>>
>> According to my test with the TDC today, this unit is not producing
>> very
>> stable data.
>> I don't have accurate pulse generator, so this is how I test the TDC:
>> 0) power the board with battery.
>> 1) use FPGA to generate time pulse:
>> reg [15:0] shift;
>> always @(posedge refclk10M) begin
>> shift <= {shift[14:0], sw_gate};
>> end
>> assign tdc_start = shift[3];
>> assign tdc_stop1 = shift[5];
>>
>> 2) use MCU to pull down sw_gate, the FPGA sync it to refclk10M domain and
>> generate input signal for TDC.
>>
>> 3) use TDC to test the time betwen tdc_start and tdc_stop1
>>
>> The result is in tdc_test.zip. number * 100ns = time between tdc_start and
>> tdc_stop1. (TDC highspeed clock is refclk10M/2).
>>
>> There 2 issues from the test:
>> 1) As we can see from the data, the number is around 1.98x not 2.00x. So
>> there is about 2ns delay between tdc_start and tdc_stop1 for this simple
>> test code. If it is from the PCB trace and something inside FPGA, this
>> part
>> should be a constant value at certain temperature. I can calculate it by
>> measuring 2 cycles and 3 cylces. My current code has not implement this
>> part, it should provide some improvement. 2ns time error for 1s gate, that
>> is something.
>> 2) For a 90ps TDC, I think the result should be something like +-0.001
>> cycle. But I get something like +-0.003 cycle. I do not know the reason
>> for
>> now.
>>
>>
>>
>>
>>
>> 2014-12-27 22:58 GMT+08:00 Bob Camp <kb8tq at n1k.org>:
>>
>> Hi
>>>
>>> (In reply to several posts. It’s easier for me this way)
>>>
>>> Ok, that’s good news !!! (and useful data)
>>>
>>> Your counter performance degraded a bit when you put in 5 db and not much
>>> when you put in 8 db.
>>>
>>> It’s also maybe *too* good news. I suspect that cross talk between the
>>> channels may be impacting your results.
>>>
>>> Next step is to try it with two independent sources and a bit more
>>> attenuation. When you try it with two sources, you need to attenuate
>>> first
>>> one source and then switch the attenuators to the other source. That will
>>> help you see if crosstalk from one channel is more of a problem than from
>>> the other channel.
>>>
>>> One parts hint:
>>>
>>> Cable TV attenuators are much cheaper than their fancy 50 ohm
>>> MIniCircuits
>>> cousins. They are also something you can pick up down at the corner
>>> electronics store. For this sort of testing they are perfectly fine to
>>> use.
>>> At this point in the testing the mismatch between 75 ohms and 50 ohms is
>>> not a big deal. You will need to adapt connectors, but you probably still
>>> will save money.
>>>
>>> =======
>>>
>>> Op-amps that have enough bandwidth and performance for a high input
>>> impedance counter input are rare items. They also are not cheap. Often
>>> they
>>> come as some sort of current feedback part with low(er) input impedance.
>>> If
>>> you want your counter to work to 300 MHz, it should accept a 300 MHz
>>> square
>>> wave. That might mean passing the third or even the fifth harmonic of the
>>> square wave. An input channel with 900 or 1500 MHz bandwidth is quite a
>>> challenge.
>>>
>>> One very simple solution is to just grab a high speed comparator like the
>>> one used by Fluke / Pendulum (ADCMP565). Drive it directly with your
>>> input
>>> or clock. Make it your front end device. That’s not an ideal solution,
>>> but
>>> it will give you the bandwidth and a reasonable input impedance. It
>>> requires messy things like a negative supply or a “fake” ground (so
>>> would
>>> the op amp). It also has an ECL output that needs to be converted to
>>> match
>>> your FPGA ( hint: use the clock inputs, they are LVPECL compatible).
>>> Driving into the FPGA with a differential signal is probably needed to
>>> reduce crosstalk.
>>>
>>> No matter how you do it, input channels are *not* an easy thing to do
>>> properly. Even on commercial counters, they often are easy to fool.
>>> Designing one is only the start. Fully testing it is equally complex.
>>>
>>> =========
>>>
>>> Do not underrate your skills in any way. You are doing far more on this
>>> project than any of the rest of the list members have done. We have
>>> talked
>>> and talked forever about these chips. We talk a lot about these ideas. We
>>> suggest lots of complex solutions to various possible problems (like the
>>> expensive comparator I suggested above). What we almost never do is
>>> actually build a counter. If we build something we don’t fully test it. I
>>> have never seen any list member share their results the way you have. I
>>> suspect that most of us (yes this includes me) are a bit to scared of the
>>> response.
>>>
>>> Please do not stop your work. Keep letting us know how it is going. This
>>> is very exciting !!!
>>>
>>> Bob
>>>
>>> On Dec 27, 2014, at 8:22 AM, Li Ang <lllaaa at gmail.com> wrote:
>>>>
>>>> Hi Bob,
>>>> Here is the data and test scheme.
>>>> It does not show much difference.
>>>>
>>>> 2014-12-26 22:12 GMT+08:00 Bob Camp <kb8tq at n1k.org>:
>>>>
>>>>
>>> _______________________________________________
>>> time-nuts mailing list -- time-nuts at febo.com
>>> To unsubscribe, go to
>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>> and follow the instructions there.
>>>
>>>
>>>
>>> _______________________________________________
>>> time-nuts mailing list -- time-nuts at febo.com
>>> To unsubscribe, go to https://www.febo.com/cgi-bin/
>>> mailman/listinfo/time-nuts
>>> and follow the instructions there.
>>>
>> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/
> mailman/listinfo/time-nuts
> and follow the instructions there.
>
More information about the Time-nuts_lists.febo.com
mailing list