Bert asked me to send an update on the FE-5680 tempco mod progress.
It appears that the FE-5680A temperature signal (or maybe it's really
a current sense signal?) can be disabled by removing a single 10k
0805 surface mount resistor.
Using Elio Corbolante's terrific high-res scans, I've noted the
resistor location: http://www.n5tnl.com/time/fe-5680a/lobotomy.png
Why would you want to disable temperature compensation? As we've
seen, the unit's firmware will adjust the DDS frequency as the
temperature signal changes. If you're using the '5680 inside a
control loop, it's likely to conflict. By removing the resistor, that
channel of the 12 bit ADC will be tied to ground through an existing
2.21k resistor. The unit will see a constant 0 counts from the ADC
and assume it's really cold.
I modified one unit and monitored it for a few hours over a range of
temps, running it nice and hot with no heatsink, then blasting it
with a fan and placing it on an ice-cold heatsink. I observed no
change in the DDS tuning words.
It's a really easy mod--remove four screws, set aside the insulator
sheet, and apply your hot leucotome/soldering iron.
I've also found a simple mod to replace the temperature signal with
the output of the unused trimpot. This allows you to simulate any
temperature you want. If there's any interest, I'll set up a test and
monitor the DDS tuning words as the unit's firmware tries to adjust
to the fake temp signal.
--
newell N5TNL