[time-nuts] Lucent KS-24361, HP/Symmetricom Z3809A, Z3810A, Z3811A, Z3812...
Robert Darby
bobdarby at triad.rr.com
Thu Oct 23 18:59:57 UTC 2014
Also of interest is the three pin connector behind the osc labeled Vref,
Rtn, and EFC. Present on both boards. Manufacturing test input?
bob darby
On 10/23/2014 1:19 PM, Bob Stewart wrote:
> My units came in today. What I got appears to be new-in-box. It's probably the only thing I'll ever get with a blue Agilent sticker on the box. =) It has a yellow Symmetricom notice inside the box.
> The circuit board appears to be the same on both units, but that says nothing about the firmware, of course. The REF-1 has an Oncore receiver labeled TM-AB - whichever one that is, small parts to support it, and a TNC connector for the GPS receiver.
>
> The REF-0 is missing everything related to the receiver, and has an SMA for the 10MHz output in the space where the REF-1 has the TNC along with a few extra small parts. This is a shared space with both SMA and TNC pads, though they don't seem to share the same electrical path. Since the SMA and TNC share the same physical space, even if the 10MHz is available somewhere, you'd have to do some surgery on the case before you could bring it out. Probably by adding a hole in the case for the GPS antenna and using the pad space for the SMA.
>
> It will be a day or two before I have the bits to apply power and connect an antenna. So, that's what I know. I'd probably just break something if I tried to find and bring out the 10MHz, so I'll have to leave that to someone else. But, the appropriate signals need to get between the boards, so I wonder what's on the Interface pins? Maybe just arbitration, 1PPS, and sawtooth comms?
> In my case, I do need the 10MHz, so I'm just as happy to have bought both units at this point. Maybe, down the road, someone will come up with the mods to convert a REF-1 into a REF-0, and vice versa, unless the firmware prevents that.
>
> Bob
> From: GandalfG8--- via time-nuts <time-nuts at febo.com>
> To: time-nuts at febo.com
> Sent: Tuesday, October 21, 2014 5:59 AM
> Subject: Re: [time-nuts] Lucent KS-24361, HP/Symmetricom Z3809A, Z3810A, Z3811A, Z3812...
>
> It seems from the auction revision table that this seller has been offering
> these for some time, so perhaps another "hidden" gem:-), but it's perhaps
> also worth noting that if this system functions on similar principles to
> earlier RFTG kit then the GPS conditioning is only applied to the unit
> actually containing the GPS module, with the other unit intended as a standby
> should the first one fail.
>
> In other words, unless the system redundancy is really required most users
> would probably only need the GPS based unit, or would at least be better
> off buying two of those for the same money that the "matched" pair would
> cost.
>
> The only advantage, as far as I'm aware anyway, of the non-GPS unit is that
> it contains a 10MHz output.
> However, Skip Withrow published modification details in January 2013
> showing how straightforward it was to add the the 10MHz output, to the
> RFTGm-II-XO module, the PCB location for the socket was already available, so I
> would suspect it wouldn't be too difficult on these either.
>
> Regards
>
> Nigel
> GM8PZR
>
>
> In a message dated 20/10/2014 05:53:29 GMT Daylight Time,
> stewart.cobb at gmail.com writes:
>
> Fellow time-nuts,
>
> This (long) post is a review of the HP/Symmetricom Z3810A (or Z3810AS)
> GPSDO system built for Lucent circa 2000. I wrote it because I looked
> for more information before I bought one, and couldn't find much.
> It's relevant because (as of this writing), you can buy a full system
> on the usual auction site for about $150 plus shipping. For those of
> you lamenting the dearth of cheap Thunderbolts, this looks like one of
> the best deals going. The description of these objects does not
> include "GPSDO", so time-nuts may have missed it. Search for one of
> the part numbers in the subject line and you should find it.
>
> So what is it? It's a dual GPSDO built by HP as a reference
> (Redundant Frequency and Time Generator, or RFTG) for a Lucent
> cell-phone base station, built to Lucent's spec KS-24361. Internally,
> it's a close cousin of a later-model Z3805A. Externally, it looks to
> be almost a drop-in replacement for the earlier RFTG system built to
> Lucent's spec KS-24019. That was a redundant system containing one
> rubidium (LPRO, in the one I have) and one OCXO in two
> almost-identical boxes. That spec went through several revisions with
> slightly different nameplates and presumably slightly different
> internals. You can generally find one or two examples on the auction
> site (search for RFTG or KS-24019).
>
> This system is similar, but the two boxes each contain a Milliren
> (MTI) 260-0624-C 5.000MHz DOCXO, and neither contains a rubidium. The
> Milliren DOXCO is the same one used in the later models of the HP
> Z3805A / 58503A. It's a very high-performance DOCXO, in the same
> class as the legendary HP 10811, and better than the one in most
> surplus Thunderbolts. The 5 MHz output is multiplied up to 10 MHz in
> at least one unit, and 15 MHz in both units. I don't have the ability
> to measure phase noise on these outputs, but I'd be interested to see
> the results if someone could.
>
> Nomenclature: The Z3810AS (there always seems to be an "S" at the
> end) is a system consisting of the Z3811A (the unit containing a GPS
> receiver), the Z3812A (the unit with no GPS receiver), and the Z3809A
> (a stupid little interconnect cable). The GPS receiver inside the
> Z3811A is a Motorola device, presumably some version of an OnCore.
> Where the Z3811A has a TNC GPS antenna input, the Z3812A has an SMA
> connector labeled "10MHz TP". That is indeed a 10 MHz output. It
> comes active as soon as power is applied to the unit, and its
> frequency follows the warmup curve of the OCXO. The two units have
> identical PCBs (stuffed slightly differently), and I have no doubt
> that someone can figure out how to add a 10 MHz output to the Z3811A
> as well.
>
> Operation: From the outside, these units are broadly similar to
> earlier units in the Lucent RFTG series. The (extremely valuable)
> website run by Didier, KO4BB, has a lot of information on those
> earlier units, much of which still applies here. The purpose of these
> units was to provide a reliable source of frequency and timing
> information to the cell-site electronics. The 15 MHz outputs from
> both units were connected to a power combiner/splitter and directed to
> various parts of the transmitter. The units negotiate with each other
> so that only one 15 MHz output is active at a time. The outputs
> labeled "RS422/1PPS" contained a 4800 baud (?) serial time code as
> well as the PPS signal, which were sent to the control computer.
>
> Power is applied to the connector labeled "+24VDC" and "P1", in
> exactly the same way as the earlier RFTG units. Apply +24V to pin 1
> and the other side of the power supply (GND or RTN) to pin 2. In
> these units, that power supply goes directly to an isolated Lucent
> DC/DC converter brick labeled "IN: DC 18-36, 1.9A". Presumably you
> can run both units with a 4-amp supply.
>
> Once you have applied power, connect the Z3809A cable between the
> jacks labeled "INTERFACE J5" on each unit. The earlier RFTG units
> used a special cable between two DE-9 connectors, and it mattered
> which end of the cable connected to which unit. The interconnect for
> these units is a high-density DE-15 connector (like a VGA plug). The
> Z3809A cable is so short that the two units need to be stacked one
> above the other, or the cable won't reach. It doesn't seem to matter
> which end of the cable goes to which unit. I don't know whether it's
> a straight-through cable, or whether you could use a VGA cable as a
> substitute.
>
> When you apply power, all the LEDs on the front panel will flash. The
> "NO GPS" light will continue flashing until you connect a GPS antenna.
> Once it sees a satellite, the light will stop flashing and remain on.
> The unit will conduct a self-survey for several hours. Eventually, if
> all is well, the Z3812A ("REF 0" on its front panel) will show one
> green "ON" light and the Z3811A ("REF 1") will show one yellow "STBY"
> light. This means that the Z3812A is actually transmitting its 15MHz
> output, and the other one is silently waiting to take over if it
> fails.
>
> Most time-nuts want to see more than a pretty green light. The old
> RFTG series allowed you to hook up a PC to the "RS422/PPS" port and
> peek under the hood with a diagnostic program. The program is
> available on the KO4BB website. It is written for an old version of
> Windows, and I had no luck getting it to run under Windows 7. It does
> run under WINE (the Windows emulator for Linux) on Ubuntu 12.04 LTS.
> To use it, you need to make an adapter cable to connect the oddball
> RS-422 pinout to a conventional PC RS-232 pinout. The adapter cable
> looks like this:
>
> RFTG PC
>
> DE-9P DE-9S
>
> 7 <----------> 5
>
> 8 <----------> 3
>
> 9 <----------> 2
>
> (According to the official specs, this is cheating, because you're
> connecting the negative side of the differential RS-422 signals to the
> RS-232, and ignoring the positive side of the differential signals.
> However, it's a standard hack, and it's worked every time I've tried
> it.)
>
> With that adapter, you can see the periodic timetag reports from the
> unit. The RFTG program will interpret these timetags when it starts
> up in "normal mode". However, when I try to use any of the diagnostic
> features built into the program, it crashes WINE. The timetag output
> was required for compatibility, but I suspect that HP didn't bother to
> implement the Lucent diagnostics.
>
> Instead, they added a connector which is not on the previous RFTG
> series. That connector is labeled, logically enough, "J8-DIAGNOSTIC".
> It too is wired with RS-422, so you need to use the same adapter cable
> as before. Once you do, you'll find that this connector speaks the
> usual HP SCPI command set (Hooray!). I used the official SATSTAT
> program (again under WINE on 12.04 LTS), but I'm sure that other
> programs written for this command set will work as well. The default
> SATSTAT serial port settings of 9600-8-N-1 worked for me.
>
> After about 24 hours, with a poorly-sited indoor GPS antenna, my
> system has converged to TFOM=3, FFOM=0 (the best possible numbers),
> and a "predicted 24-hour holdover uncertainty" of 5.2 microseconds,
> which is not too shabby. It found the correct day and year without
> any assistance, so if it has a "GPS week number rollover" problem,
> it's still in the future. I don't currently have the ability to
> compare the 10 MHz output to anything else. Again, if someone else
> can, I'd be interested to see the results.
>
> Additional Notes: The parts on the boards all have date codes of 1998
> or 1999. The Motorola GPS receiver has a firmware label that reads
> "02/04/00". The SCPI error logs inside the HP units were virgin when
> I first got them. They had 84 and 94 power cycles, respectively.
> Before the GPS receiver acquired time, the error log timestamps read
> "2000-05-09 00:00:00", which I interpret as a firmware release date.
>
> The PCB has an interesting feature. Next to each soldered-in pin of
> the Milliren OCXO is a single-pin socket soldered into the board. I'm
> guessing this was used in manufacturing, to temporarily install a
> Milliren and confirm that the system worked before permanently
> soldering it in. (At production prices, the Milliren would have cost
> far more than the rest of the PCB.) You might be able to use this in
> reverse, if you have a set of Millirens to test from another source.
>
> The Z3809A interconnect cable has three of the 15 pins on each end
> clipped a bit shorter than the rest. Not so short that they won't
> eventually make contact, but short enough to make contact later than
> the rest. Don't know why, but it's clearly deliberate. A lot of
> hot-plug connectors are built that way, including USB connectors. I
> have no idea what the pinout of the interconnect is.
>
> The redundant system slaves both DOCXOs to the same GPS reference.
> Inside the GPS loop bandwidth, the two oscillators will have almost
> the same frequency and will differ only by phase noise and short-term
> stability. This is almost a perfect setup for experimenting with
> certain kinds of time-nut measurements, assuming someone can figure
> out how to get 10MHz out of the Z3811A unit. If you then command both
> units into holdover, you could measure longer-term stability as well.
>
> The units are described as "new in factory sealed box". After an
> archeological investigation of the various strata of labels and tape
> on the boxes, I would say that's probably accurate. My set seems to
> have been shipped from the Agilent factory in Korea to Symmetricom in
> Sunnyvale, CA sometime in August, 2000, shortly after it was built,
> and remained untouched until I opened it. I'm guessing it was built
> and saved as part of a spares program for Lucent, and kept until
> Lucent decided they didn't need spares any more.
>
> I have no connection with the current seller of these units (or any
> other sellers, for that matter) except as a satisfied customer. I
> think I'll order another set as a spare, before the feeding frenzy
> hits.
>
> Request for help: Both the SatStat and RFTG programs run under WINE
> on stock Ubuntu 12.04 LTS (32-bit) without any tricks or special
> configuration. Neither seems to run under WINE on Ubuntu 14.04 LTS
> (64-bit). I am a WINE novice. Any hints from WINE experts would be
> appreciated. Also, I've been able to run TimeLab under WINE, but I
> can't connect it to my USB-to-488 interface, so I can't take data. If
> anyone can tell me how to set that up, I'd be extremely grateful.
>
> Cheers!
> --Stu
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
More information about the Time-nuts_lists.febo.com
mailing list