Hi,

We have already hacked a DC-DC converter to feed the output board with 
separate power. It was relatively easy.

Haven't had the time too look at doing the same to the external trimmer.

Cheers,
Magnus


On 09/30/2014 02:09 AM, Björn wrote:
> Hi Chris,
>
> Nice work!
>
> I can't help you with the faulty unit. There are others on the list with deeper insight in that OSA Cs core. Hope they have time to chime in.
>
> But I do have a telco variant of the same Cs core running in the basement. Its running continously (ocxo and ion pump) but its not burning the last fumes of Cs in the tube. This is since I don't see access or money available for a replacement tube.
>
> As you noted. With ocxo warm and cosy and the ion pump running it does not take long for Cs lock when needed.
>
> A mode where you can use the unit as premium ocxo (bva :-) ) standard that can be calibrated against the cs tube periodically as needed would suit my situation.
>
> However there are (at least) two mods needed. Because unless the cs tube burnes, there is no power to the
>
>   1)  output buffer card. And thus no output signal.
>   2) potentiometer adjusting the bva frequency. Thus without cs tube running the bva is off freq.
>
> This design might have been a safe one in the original market.
>
> But I think modding the unit to let it run mostly as a very stable crystal standard is better for hobbyist use.
>
> Have others done these kind of mods to prolong the Cs tube life?
>
> Kind regards,
>
>            Björn
>
>
>
> <div>-------- Originalmeddelande --------</div><div>Från: Chris <syseng.greenfield at btconnect.com> </div><div>Datum:2014-09-29  22:43  (GMT+01:00) </div><div>Till: time-nuts at febo.com </div><div>Rubrik: [time-nuts]  Oscilloquartz 3210 Cesium Standard </div><div>
> </div>Another update on progress:
> --------------------------
> Now at the stage where two of the three 3210's are working. The fault
> that was blowing the line fuse turned out to be a shorted reservoir cap.
> Managed to find an Ebay seller in Greece with exactly the same voltage,
> capacity and outline, though different manufacturer. The caps have a
> stud mount at the base of the can, so not easy to find, but bought four
> at just over 3 ukp each and replaced both caps in the psu, leaving two
> for spares.
>
>  From cold, the spec is up to 90 minutes warmup time, but if the OCXO
> and pump have been running for some time, both units lock up within 4
> minutes. Have done some setup. For example, normalised the gains and
> offsets in the preamps as per the manual and both units show the
> expected preamp level of ~165mV and 2nd harmonic amplitude of 9-10 on
> the meter, so assume both tubes are in good condition.  None of the
> settings were very far out.
>
> Took a set of measurements from one of the good units and the faulty for
> comparison, with similar results, Good unit as follows, but faulty unit
> more or less the same:
>
> Ion pump = EHT ok, meter = 0
> Electron multiplier EHT voltage = 1800 volts, (100Mohm hv probe on o/p
> wire)
> Cesium oven = Not open circuit, voltage = 5.5 volts
> Ioniser = not open circuit, 1 volt p-p square wave at ~26KHz
> Synthesiser output = 12.6317715, follows variation in OCXO and in lock
> Backplane test points, tpb, tpc = 8.2 volts p-p square wave, ~137 Hz.
> Multiplier 12.6... MHz input = 2v p-p (Scope)
> Multiplier 137Hz input = 2.25v p-p (Scope)
> Multiplier 180MHz output = +26DBm (HP 3406A rf voltmeter + 20DB attenuator)
> Microwave tap on waveguide = -14.5DBm, threaded attenuator works
> (HP432A, 478A)
>
> The 3406 and 432 are quite old and the sensor head is even older, but
> should be reasonably accurate. The 478A head was coupled to the
> microwave tap using a ~3 inch length of rigid coax, so not much
> attenuation. Hopefully, these figures may help others trying to debug
> these standards, but if there's anything i've missed, please let me know.
>
> For the faulty unit, even with all the levels as expected, there is
> still no signal, nor 2nd harmonic, even with the preamp gain turned up
> to max. Connected a Fluke electrometer in series with the preamp input
> socket, but the current is at least 100 times down on 1nA. Looks like
> the tube is completely lifed, or has an internal fault. /FX:  Have
> visions of a sad, faithful and slowly dying 3210 left for decades in a
> rack,  in the dark, powered up, but long forgotten after the accurate
> reference it provided ceased to be used anywhere in the organisation. A
> ship adrift for thousands of years, but all systems still at least
> partially working and waiting to be discovered :-).
>
> So what else am I missing in this puzzle ?...
>
> Regards,
>
> Chris
>
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>



More information about the Time-nuts_lists.febo.com mailing list