[time-nuts] 53230A noise floor
Poul-Henning Kamp
phk at phk.freebsd.dk
Fri Apr 3 15:06:56 UTC 2015
--------
In message <CAPnVNRUdeRCi4suLHB9dJJF7sQwiXa5oVaRjWTEYLpoht+MyUA at mail.gmail.com>
, Anders Wallin writes:
>I ran some tests on a new 53230A counter that just arrived:
>http://www.anderswallin.net/2015/04/keysight-53230a-noise-floor-test/
Anders,
I think that setup didn't measure what you think.
When you feed the counter EXT REF from the same source you measure,
all the transitions will happen at a certain point relative to the
internal clockcycles of the counter, and therefore be subject to
only one of many possible noise-situations.
A much better way:
Use the "good" signal for your CH1/CH2 inputs, and also feed it to a
slightly offset synthesizer generator for the EXT REF.
If you feed the EXT REF 10.0001 MHz (relative to CH1/CH2, the absolute
value is not really important) then over a period of 10000 seconds
your test-setup will sweep the internal clock-cycle of the counter
at least once.
Then plot the MIN/MAX/STDDEV of the measurement versus time and you
will discover that the noise floor is not even close to being even.
If you find a particular good spot and want to use it for measurements,
the way to go is to tune the length of the cable for EXT REF (or use
a tweakable delay line)
--
Poul-Henning Kamp | UNIX since Zilog Zeus 3.20
phk at FreeBSD.ORG | TCP/IP since RFC 956
FreeBSD committer | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.
More information about the Time-nuts_lists.febo.com