[time-nuts] Once again about counter calibration
Charles Steinmetz
csteinmetz at yandex.com
Sun Apr 26 06:01:01 UTC 2015
>Using GPSDO 10 Mhz as REF signal, I was able to calibrate OCXO. And
>now its potentiometer position was nether at its both extremes. The
>reading on 5386a (using 10 sec gate) fluctuate from 9.999999993M to
>10.000000007M. Again, may be I need to wait much longer when OCXO
>will be stable.
So, you are using a 5386A with TCXO to meaasure an OCXO, and the
reading you get drifts 14 e-10 (or 1.4 e-9), nicely centered on 10
MHz. This error band includes the error of both the TCXO and the
OCXO (presumably, the TCXO contributes most of this error, but you
don't have enough information to conclude that yet). [This error
also includes the jitter in the trigger circuitry of the 5386A, but
if you are using a 10 second gate time that should be small compared
to the drift of the TCXO.]
I don't believe you can reasonably expect this setup to exhibit less
drift than what you are seeing. In fact, I'm surprised the drift
range is as low as 1.4 e-9. Waiting will probably change the center
of the drift range (it will not stay symmetrically +/- 7 e-10 above
and below 10.000000000), but not the magnitude of the range (14 e-10
peak to peak).
>So, I think the best approach will be using 10Mhz GPSDO as ref.
>signal for this counter. In another case, I'll need to wait to warm
>it up (the manual advised only 30 minutes. But I am not sure).
Generally, a time nut would let a counter warm up for 24 to 48 hours
before doing any critical work. And yes, the GPSDO is probably the
most stable of the three oscillators you have (I assume you leave it
running 24/7/365), so the best you will be able to do is to feed it
into the 5386A's external reference input and measure the OCXO. You
should see a smaller drift range than the 1.4 e-9 you are seeing now,
which will indicate that most of the drift you are seeing now is due
to the TCXO (and confirm the expectation noted above).
Most time-nuts use a "distribution amplifier" to supply separately
buffered copies of their best oscillator to various instruments and
other equipment. In your case, you would feed the GPSDO into the
distribution amp and then use the distribution amp's multiple outputs
to feed the 5386A's reference input, your various radios, etc., etc.
Best regards,
Charles
More information about the Time-nuts_lists.febo.com