[time-nuts] Measuring frequency rather than tuning crystal
Dave Martindale
dave.martindale at gmail.com
Fri Feb 27 21:23:19 UTC 2015
The TF930/960 does have a calibration procedure that is performed from
the front panel. Basically, you feed it a stable input from any known
source (so both 1 Hz and 10 MHz from a GPSDO should work) and then
adjust until the displayed frequency agrees with the known input
frequency. The resolution of this setting is quite a bit better than
the stability of the TCXO in the box.
Now, this process could occur either by actually adjusting the frequency
of a VCTCXO in the box using a DAC, or by changing a calibration
constant stored in the memory of the device. I suspect it's actually
the former, because the instructions say that the adjustment path has a
low-pass filter that you need to allow to settle. This wouldn't be
necessary if the calibration simply changed a stored number.
Dave
On 27/02/2015 15:08, Paul Alfille wrote:
> I don't think your TTi TF930 has a GPS input to calibrate against, based on
> a quick perusal of the data sheet. I would guess that the calibration
> constants are thus fixed from the factory (including temperature
> coefficients).
>
>
> On Thu, Feb 26, 2015 at 3:36 PM, James via time-nuts <time-nuts at febo.com>
> wrote:
>
>> I presume that this is what my TTi TF930 does. Calibration is closed box
>> so I guess the TCXO is free running and the micro inside just uses
>> calibration constants.
>>
>> James
>>
>>
>>
>>
>>
>>
>>
>>
>> -----Original Message-----
>> From: Paul Alfille <paul.alfille at gmail.com>
>> To: Discussion of precise time and frequency measurement <
>> time-nuts at febo.com>
>> Sent: Thu, 26 Feb 2015 19:02
>> Subject: [time-nuts] Measuring frequency rather than tuning crystal
>>
>>
>> I have a couple of HP 5370s with the beaglebone brain transplant. They
>> come
>> with a nice 10811 that has a little adjustment screw.
>>
>> Testing against a
>> Thunderbolt or KS-24361 the 5370 is off by less than 1Hz.
>>
>> I know the
>> traditional method would be to adjust the crystal slowly and
>> make careful
>> measurements, but since I have a fancy computer in there, I
>> wonder if I could
>> just adjust the frequency in software. 64-bit floating
>> point numbers should
>> have sufficient accuracy. All reported measurments
>> would be corrected for the
>> actual reference frequency.
>>
>> Basically, I'd have a 10000000.226 Hz internal
>> reference.
>>
>> In fact, could I connect the beaglebone to a a GPS 1 pps source
>> and make
>> this a GPS-disciplined-software-corrected oscillator.
>>
>> So my
>> question is is this a known technique? The discipline feedback
>> circuit seems a
>> little different, I'd adjusting for drift and offset, but
>> not the gain of
>> control-oscillator
>> linkage.
>> _______________________________________________
>> time-nuts mailing
>> list -- time-nuts at febo.com
>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the
>> instructions there.
>>
>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts at febo.com
>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>>
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
More information about the Time-nuts_lists.febo.com
mailing list