[time-nuts] PN measurement of SSINE to CMOS converters
Bruce Griffiths
bruce.griffiths at xtra.co.nz
Mon Oct 26 02:19:14 UTC 2015
The technique I used to measure the PN contribution of the LTC6957-4 at
10Mhz
Low noise 10MHz sinewave OCXO (Trimble 37265 removed from Thunderbolt
(Thunderbolt has rather noisy output amplifier)). OCXO PN measured
against a pair of similar OCXOs via correlation using the TimepodConnected
to low noise RF amplifier (push-pull transformer feedback CB Norton
amplifier NF ~ 2dB Z10043 from Clifton Laboratories) with a gain of
~12dB.Output (~19dBm) from amplifier is split by a minicircuits 1:2 splitter
(Minicircuits ZFSC-2-4)3dB 50 ohm passive attenuator reduces one output
to ~ 14dBm this is connected to the Timepod reference input.
The other output from the splitter is attenuated with a passive 50 ohm
attenuator to the desired input signal level for the DUT.
The DUT output drives a minicircuits 1:2 splitter via a lowpass filter.Each of
the splitter outputs is connected to either Timepod Ch0 or CH2 via an RF
amp (in my case a HELA10-D) with a gain of around 10dB.
Cross correlation ensures that the relatively high noise of the HELA10-D's
isn't too significant, the trade off being that more time is required to
average down their contribution to the measured noise.
I use an RF power meter to measure the output of each HELA10-Dto
ensure that the Timepod input signal doesn't exceed the maximum rated
input for the Timepod.
Alternatively if one has a sufficiently low noise amp like the Q-bit QB188 (NF
~3dB Gain ~ 15 dB) a single amp connected directly to the Timepod signal
input via a passive attenuator should suffice.
For most useful comparators or CMOS gates an ouput circuit like that on
the LTC6957-4 evaluation board should suffice.
Bruce