[time-nuts] Unified VCXO Carrier Board
Gerhard Hoffmann
dk4xp at arcor.de
Tue Oct 27 17:03:27 UTC 2015
Am 27.10.2015 um 13:50 schrieb Charles Steinmetz:
> Gerhard wrote:
>
>> Since the LM329 was mentioned above, I tried it also. Another surprise.
>> HOW CAN THEY MAKE SUCH A NOISY SUBSURFACE ZENER???
>>
>> 0 dB = 1nV/sqrt Hz
>
> The datasheet for both NS and LT LM329s shows broadband noise of
> <70nV/sqrtHz, with a corner at ~100Hz, rising to a touch over
> 100nV/sqrtHz at 10Hz. That is consistent with the LM329s I have used.
> So, according to the datasheet, the 10Hz noise ought to be a touch
> above 40dB on your graph and the curve should level off above 100Hz at
> about 30dB -- if I have done my dBs right. Still somewhat higher than
> most of the other diodes you tested, but not outlandishly so.
>
> Your data seems to be *much* worse than this, with a 1/f noise corner
> at ~10kHz (!!), rising to what appears to be about 1mV/sqrtHz (!!) at
> 10Hz. Is it possible you got fakes? (I know someone who
It's bad, but not that bad. 10 times the voltage is 20 dB more. So
0dB = 1nV/sqrt Hz
20 dB = 10 nV
40 dB = 100 nV
60 dB = 1uV
80 dB = 10 uV/sqrt Hz
The wideband noise at, say, 46 dB would be 200 nV/sqrt Hz versus the
claimed 75 nV typical
but the corner frequency is really bad. I bought them last year from
Digi-key, still with
National as manufacturer on the bag, not TI.
I'll test a few more next weekend.
regards, Gerhard
-------------- next part --------------
A non-text attachment was scrubbed...
Name: lm329density.png
Type: image/png
Size: 34577 bytes
Desc: not available
URL: <http://febo.com/pipermail/time-nuts_lists.febo.com/attachments/20151027/cf29ebed/attachment.png>
More information about the Time-nuts_lists.febo.com
mailing list