[time-nuts] State of the art of crystal oscillator measurements
KA2WEU at aol.com
KA2WEU at aol.com
Thu Aug 11 00:04:51 UTC 2016
The _time-nuts at febo.com_ (mailto:time-nuts at febo.com) reformatted the
offset / phase noise table so it became almost unreadable compared to the
original , sorry,
Ulrich
In a message dated 8/10/2016 8:00:28 P.M. Eastern Daylight Time,
time-nuts at febo.com writes:
Test Equipment Improvement
____________________________________
The measurement of the 100 MHz XO was done on an improved R&S FSWPs.
Measured data is in the table below:
XO 100 MHz
FSWP SN 101143
FSWP SN 101192
Offset
Phase Noise in dBc/Hz
0.1 Hz
-45.24
-49.11
1 Hz
-82.30
-82.33
10 Hz
-113.87
-113.74
100 Hz
-143.27
-141.61
1 kHz
-166.55
-167.26
10 kHz
-180.54
-186.91
100 kHz
-186.41
-195.43
_______________________________________________
time-nuts mailing list -- time-nuts at febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
More information about the Time-nuts_lists.febo.com
mailing list