[time-nuts] State of the art of crystal oscillator measurements

Mike Feher mfeher at eozinc.com
Thu Aug 11 19:51:29 UTC 2016


kT is indeed relevant for a physical implementation. - Mike 

Mike B. Feher, EOZ Inc.
89 Arnold Blvd.
Howell, NJ, 07731
732-886-5960 office
908-902-3831 cell

-----Original Message-----
From: time-nuts [mailto:time-nuts-bounces at febo.com] On Behalf Of Richard (Rick) Karlquist
Sent: Thursday, August 11, 2016 1:40 PM
To: Discussion of precise time and frequency measurement; KA2WEU at aol.com; tvb at leapsecond.com
Cc: enrico.rubiola at gmail.com
Subject: Re: [time-nuts] State of the art of crystal oscillator measurements



On 8/11/2016 5:01 AM, Mike Feher wrote:

> This may be a naive question, but, how can you achieve results that 
> are so close to, and sometimes at further out are below kT? Thanks & 
> 73 - Mike

> Mike B. Feher, EOZ Inc.
>

kT per see is not the relevant parameter.  It is the ratio between kT and signal level that determines the limit on phase noise.  Also, when you extract an oscillator signal through the resonator, then at offsets from the carrier, the resonator filters out noise and the output noise can be way below kT.  Ulrich first published on this in 1977.  Earliest reference I know of.
I actually remember reading it in 1977.

The old HP608 signal generator routinely produced far out phase noise floor well below kT.  AFAIK, for this particular spec, it better than any other signal generator ever built.  What's the secret?  It has a tracking POST-selector filter that follows the oscillator when you turn the mechanical tuning knob.  There is a little tracking adjustment knob to peak it.  Not really magic.

Rick