[time-nuts] Introduction
Mark Spencer
mark at alignedsolutions.com
Wed Feb 24 16:20:47 UTC 2016
I've found with my HP5370B's they perform best over a fairly narrow range of input signal levels. (1 to 2 volts If I recall correctly ?) Some experiments might be in order for yours. Virtually all my testing was done with 5 or 10 MHz sine waves so this may or may not be helpful in your case. I found in line attenuators and an oscilloscope to be helpful in making measurements using my HP5370B's.
I've never done a full alignment of mine so this behaviour may not be typical.
All the best
Mark S
Sent from my iPhone
>> On Feb 24, 2016, at 1:36 AM, Bruce Griffiths <bruce.griffiths at xtra.co.nz> wrote:
>>
>> On Wednesday, February 24, 2016 12:55:42 AM Enrico Bellotti wrote:
>> Hello to all,
>>
>> first of all, thank you for the great and useful work that the time-nuts
>> have done over the years.
>>
>> I have finally been able to gather all my counters (HP5335A, HP51131A,
>> HP5370A, HP5370B #1, HP5370B #2) and try to do some "simple" measurements.
>>
>> I have started testing and comparing the instruments I have available using
>> the approach that was discussed a while ago on the list and also outlined
>> by John Ackermann (http://www.febo.com/pages/hp5370b/). Specifically, the
>> Adev for a time interval measurement on a (90ft of) RG58A/U cable.
>>
>> I have attached a PDF file with the results and some additional details of
>> the test setup.
>>
>> From what I have understood the Adev at one second is related to the
>> counter resolution. The results I have obtained seem to be reasonable
>> except for HP5370B #2. This instrument seems to be marginal at best. Does
>> anybody know if the measured value for HP5370B #2 is a symptom of a
>> multifunction or simple need for calibration?
>>
>> Thank you for any comments/suggestions/corrections you may have.
>>
>> Best.
>> Enrico
> Try repeating the test with a low noise external reference/If that doesn't
> improve the results try realigning the filters etc in the 10-200MHz multiplier.
>
> Bruce
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
More information about the Time-nuts_lists.febo.com
mailing list