[time-nuts] Valpey-Fisher VFT22H OCXO Output
Bob Camp
kb8tq at n1k.org
Sat Jan 30 14:33:45 UTC 2016
Hi
Unless it’s a really weird termination impedance, it’s broke. The SMA is a pretty
good indication that it *should* work into a 50 ohm load.
The only other possibility is that your probe is taking the load far enough off of 50 ohms
to mis-tune the output filter. I’d set the scope to 50 ohm input and cabe the output directly to it.
If that’s not practical, put a 6 db 50 ohm pad on the output of the oscillator and see what it
looks like after that.
Bob
> On Jan 29, 2016, at 10:19 PM, Logan Cummings <logan.cummings at gmail.com> wrote:
>
> Hi All,
>
> Curious what could give this waveform (attached) - I presume this is
> not correct output for this oscillator (surplus/salvage from Harris
> Constellation receiver).
>
> Frequency looks OK with +2V EFC but output looks like neither sine nor
> any CMOS I've seen into 10Mohm (first capture, scope probe alone) or 50ohm
> (50ohm term at SMA connector, probed by same scope probe) .
>
> What gives? If I open up the can, what am I looking to possibly
> replace/eliminate? How to open it? It's soldered shut (2x2in case).
>
> I need a good low-noise 100MHz sine wave for my application so I'm
> tempted to open this guy up and replace the output circuitry with something
> more suitable.
>
> Thanks in advance for any pointers/info on these units!
> -Logan
> <tek00000.png><tek00001.png>_______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
More information about the Time-nuts_lists.febo.com
mailing list