[time-nuts] uncertainty calculations
jimlux
jimlux at earthlink.net
Fri Apr 14 15:49:07 UTC 2017
On 4/14/17 8:37 AM, jimlux wrote:
> If one is counting an unknown 1pps source with a counter that runs at 10
> MHz (e.g. the error in any one measurement is uniformly distributed over
> 1 ppm) and you collect 100 samples,
> is the (1 sigma) measurement uncertainty 0.1ppm * sqrt(100)/sqrt(12)
>
> (standard deviation of a uniform distribution is 1/sqrt(12) )
>
> (assuming for the moment that both sources have no underlying
> variability - we're talking about the *measurement uncertainty*)
>
Oops..
0.1 ppm * 1/sqrt(N) * 1/sqrt(12)
That is, the standard deviation goes down as sqrt(N)
More information about the Time-nuts_lists.febo.com