[time-nuts] HP 5372A channel c option 030
Magnus Danielson
magnus at rubidium.dyndns.org
Sat Jan 7 02:12:56 UTC 2017
Hi Bob,
There is also the A20 board for the C-channel consisting on a series of
attenuator, amplifiers/limiters and a divide by 4 before reaching that
SMA connector. You then flip the option switching on the motherboard.
I got the C-channel on mine, but you should be looking into
05372-90016.pdf which is the service manual, chapter 12.
If you have a replacement board that does the same operation, you can
enable the C-channel without too much effort as you install it.
Cheers,
Magnus
On 01/07/2017 02:26 AM, Bob Bownes wrote:
>
> While I had the 5372a apart today to replace the battery, I noticed that there is an SMA for channel C on the events board and diagram showing a cable from there to the front panel.
>
> This suggests to me that the upgrade is a cable and a key, jumper, or memory setting.
>
> Anyone ever done the upgrade?
>
> Bob
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
More information about the Time-nuts_lists.febo.com
mailing list