[time-nuts] J06 HP-59992A time interval calibrator for HP-531xxcounters
Bob kb8tq
kb8tq at n1k.org
Sat Jul 8 23:18:49 UTC 2017
Hi
Good means whatever the 5313x needs for calibration. If that is four signals that are
crossing zero within < 10 ps of the “correct time” then that is the definition of good in this case.
Rise time delay, fall time delay are rarely the same in logic gates. Propagation inside a chip to
point A may well be different by nanoseconds relative to the propagation to a very similar
point B. All of that would mess up a signal that *might* need to be 50/50 to within 10 ps or
a second signal that must cross zero half way in-between (also to within 10 ps).
If you want to have a lot of fun with this, pull out the timing analysis tool for your favorite
FPGA and start fiddling around.
Bob
> On Jul 8, 2017, at 5:53 PM, Hal Murray <hmurray at megapathdsl.net> wrote:
>
>
> tvb at LeapSecond.com said:
>> The PIC dividers are good to a couple ps. I suspect the larger issue is the
>> PCB and wiring design.
>
> What does "good" mean?
>
> I'd expect the variations due to power or temperature would be easy to
> measure.
>
> Delay through classic CMOS is linear with absolute temperature and inverse
> linear with supply voltage.
>
> The classic way to get time-nuts level noise on FPGA outputs is to wiggle a
> nearby pin. That shouldn't be a problem with a dedicated PIC but would
> probably show up if you are generating multiple frequencies.
>
>
> --
> These are my opinions. I hate spam.
>
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
More information about the Time-nuts_lists.febo.com
mailing list