[time-nuts] 33120A purchase off epay
jimlux
jimlux at earthlink.net
Mon Mar 6 15:22:50 UTC 2017
On 3/5/17 9:14 PM, Christopher Hoover wrote:
> My HP 33120A Function Generator died. (I'm still trying to fix it, but it
> isn't obvious what's wrong ... rails check so I'm suspicious the EEPROMs
> gave up the ghost.)
>
> I bought a new one on epay week befor last for a fair price.
>
> It arrived and I was surprised/excited to find it had the external
> reference option. Just what a TimeNut needs.
>
> The external ref option wasn't mentioned explicitly in the listing,
> although I can see it (in retrospect) in one of the pictures.
>
I don't know about on the 33120, but on the 33620 series, the external
reference doesn't actually phase lock the internal clock for the ARB.
It's more of a frequency reference, and the output is steered to match
the frequency of the external reference. (counters and NCOs) The phase
noise of the 36620 is entirely determined by the internal oscillator,
and the ADEV is a bit tricky - it's like the ADEV of a disciplined
oscillator, not a phase locked loop.
There are ways to sync multiple 33620 and 33120s (and the earlier 3325)
so that you have multiple outputs with a phase coherent relationship.
You have to do it manually.. feed them all the same reference, then
adjust til they're synced, then they'll stay in sync.
Here's Agilent's ap note on externally syncing 33120s
http://cp.literature.agilent.com/litweb/pdf/5988-8151EN.pdf
about your option 001
http://www.keysight.com/upload/cmc_upload/All/33120A_Opt001.pdf?&cc=US&lc=eng
service manual and schematic of your Opt001
http://literature.cdn.keysight.com/litweb/pdf/33120-90022.pdf
(From Keysight about the 33620 series:
When using an external 10 MHz time base with this instrument, a PLL
syncs together the internal oscillator with that external signal, as you
have surmised. To my understanding, because of this however, the
stability of the function generator’s time base is still dependent on
the internal oscillator. If an OCXO is being used in the 33600, the
instrument will be limited to a input frequency range of 10 MHz ± 1Hz.
Therefore, even if an external input signal at 10 MHz was stable to 10
ppb, for instance, the instrument would still be limited to a 100 ppb
stability because of its’ own internal oscillator. This means feeding a
more stable reference signal into the function generator, will likely
not improve performance. Again, this is to my own understanding of how
this instrument works and I am going to double check this with our
factory to see if they agree or disagree. I will update you when I have
that confirmation.
)