[time-nuts] HP5065A C-field mods and optical unit mods
timeok at timeok.it
timeok at timeok.it
Tue Nov 21 12:41:01 UTC 2017
Hi Charles,
what you refer is the resistance relation referred to the LM299 source around 14VDc but you have to consider
the tuning range of the pot, in the original HP5065A is 2E10-9.
Of course, it depends on the setting value of the variable resistance, but you have to consider a possible variation of 20PPM/c on 2E10-9.
Regards
Luciano
timeok
Da "time-nuts" time-nuts-bounces at febo.com
A time-nuts at febo.com
Cc
Data Tue, 21 Nov 2017 06:31:25 -0500
Oggetto Re: [time-nuts] HP5065A C-field mods and optical unit mods
Luciano wrote:
> about the C field ten turn pot you write is a low TC pot. Normally a good Vishay pot is 20 PPM/C, This is in conflict with the 1PPM resistors ad voltage reference.
In the schematic Corby posted, the potentiometer (used as a variable
resistor) makes up a very small part of the total resistance of the
series string, so the overall tempco is dominated by the tempcos of the
fixed resistors (2.4k to 4.4k 1ppm/ C, depending on trim). The pot will
increase the overall tempco a little bit, but not anywhere near 20ppm.
Best regards,
Charles
_______________________________________________
time-nuts mailing list -- time-nuts at febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
More information about the Time-nuts_lists.febo.com
mailing list