I've used a few Pt100 RTDs over the years for gas flow temperature 
control in the 200-300C range.  At 0.385 ohms per K, the challenge of 
managing the noise is a lot tougher than a cheapo NTC thermistor, but I 
needed repeatability and stability. 
http://www.farnell.com/datasheets/2278399.pdf

I use a 4-wire Kelvin connection, with 1 mA current drive. £14 at 10-off

Neil


On 05/04/2018 16:27, Dr. David Kirkby wrote:
>
> Keysights nano volt / micro ohm meter takes
>
> https://www.keysight.com/en/pd-1000001296%3Aepsg%3Apro-pn-34420A/micro-ohm-meter?cc=ZA&lc=eng
>
> can do direct SPRT, RTD, Thermistor, and Thermocouple measurements
>
> SPRT = Standard Platinum Resistance Thermometer
>
> I don't claim to know much about this, but the uncertainty quoted for the
> SPRT probes with that meter is 0.003 deg C.
>
> When I Google SPRT probes, I see they are incredibly expensive - many
> thousands of USD each.
>
> I'm a bit puzzled there seem to be a number of 3-wire platinum resistance
> thermometers. I can understand using 4-wires for a Kelvin connection, but
> can't understand the use of 3-wires.
>
> Of course, for environmental monitor in a lab, one is most unlikely to need
> very high accuracy.
>
> Dave
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.