[time-nuts] Time interval measurement vs dual mixer method
Bob kb8tq
kb8tq at n1k.org
Sun Jan 21 14:04:36 UTC 2018
Hi
What level of stability are you trying to measure?
What sort of offset frequency are you running?
What kind (phase noise / spurs / adev) offset oscillator are you using?
What sort of limiter are you running with what sort of pre filtering?
We have run around on a lot of generalities. You may have some issues
that are specific to your setup. There is no “one size fits all” approach here.
If you are running a couple of oscillators it the THz region and trying to look
at ADEV from 1us to two months tau with one setup, a lot of what we have
been saying simply does not apply.
Bob
> On Jan 21, 2018, at 4:23 AM, timeok at timeok.it wrote:
>
>
> Corby,
> thanks for the information.
>
> I'm definitely interested in the short therm ADEV but especially for Tau from 1 day and over to test high stability standard over the time.
> I will have further tests.
> Cheers,
> Luciano
>
>
> Da "time-nuts" time-nuts-bounces at febo.com
> A time-nuts at febo.com
> Cc
> Data Sat, 20 Jan 2018 16:14:10 -0800
> Oggetto [time-nuts] Time interval measurement vs dual mixer method
> Luciano,
>
> I don't know of a commercial version.
>
> As you have seen as the two signals move away from close phase
> coincidence the system noise level will increase.
>
> This is because the common offset oscillator noise will only cancel when
> the phases are closely matching.
>
> This is mainly important for the lower Tau (like 1 to 10 seconds).
>
> I will normally adjust the phase of my reference or DUT so that myTIC is
> reading 0.00000XX (close to phase match) and very slowly rising. (with my
> setup most DUT will age downward).
>
> Then I start logging the data.
>
> A typical run with a 1 week aged quartz will show the TIC count slowly
> increase (and may wrap) and then as the aging continues the count will
> reverse and eventually go "below" zero and wrap.
>
> A phase plot will show a nice parabolic looking curve going up and then
> as the aging crosses it will curve down.
>
> If there are wraps the plotter program can take them out.
>
> Any long plot of Quartz WILL wrap!
>
> Now if you plot the AD you will see a normal type plot for a good Quartz.
>
> If the TIC count increased well past coincidence only the Higher Taus
> will be accurate.
>
> For the lower Tau you need to cut off the data past where it climbed too
> high.
>
> I typically will setup as described and run a short log of say 5 minutes.
>
> I might only keep the first 100 Seconds and plot the AD against that.
>
> This will give you an accurate plot for the lower Tau.
>
> These two plots can then be combined to give you the whole range.
>
> The attached plot of an FE405B illustrates this.
>
> The Red plot is against a very good FTS 1200 (2X10-13th at 1 thru just
> past 10 Sec)
>
> The blue continuation is against a very good HP 5065A (1.5X10-13th at 100
> Sec)
>
> So if the lower Tau are important only use logged data that are in near
> phase match!
>
> Hope this helps!
>
> Cheers,
>
> Corby
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
More information about the Time-nuts_lists.febo.com
mailing list