Hi Adrian,
as you know, to get real data up to 100K sec range, you should acquire at least 500K 1M sec.
I say this from my personal experience. Soon I will have my 30 days run for both the A and B version of the HP5065.
Very interesting the response of the GPSDO in the range 1 to 100 sec that I had never had the chance to see for lack of an instrument as your A7.
ciao,
Luciano
Da "time-nuts" time-nuts-bounces at febo.com
A time-nuts at febo.com
Cc
Data Fri, 8 Jun 2018 15:46:50 +0200
Oggetto Re: [time-nuts] Raw phase data of super-5065
This is a 6 days run of a modified 5065'B' against GPS.
Adrian
Am 08.06.2018 um 14:05 schrieb timeok at timeok.it:
> Here a ten days simultaneously run of the A and B version compared with my HP Z3816A GPSDO.
> Luciano
>
>
> Da "time-nuts" time-nuts-bounces at febo.com
> A kb8tq at n1k.org
> Cc "time-nuts" time-nuts at febo.com
> Data Wed, 6 Jun 2018 08:18:27 +0200
> Oggetto Re: [time-nuts] Raw phase data of super-5065
>
> Hi,
>
> certainly the environmental effects are significant and it is precisely for this reason that
> acquisitions have been made simultaneously on two 5065, one modified and the other not.
> To verify the results the same measures were taken with another pair of 5065, one modified and the other not.
> This second test was carried out at another site 300Km from the first one.
> Both tests showed a better and more consistent medium/long term ADEV than the unmodified HP5065A.
> I want to clarify that these measures have the objective of verifying the medium / long term stability and not the short term stability of the Super version I agree it is improved..
> Luciano
>
>
> Da "Bob kb8tq" kb8tq at n1k.org
> A "Luciano Paramithiotti" timeok at timeok.it,"Discussion of precise time and frequency measurement" time-nuts at febo.com
> Cc
> Data Tue, 5 Jun 2018 10:05:34 -0400
> Oggetto Re: [time-nuts] Raw phase data of super-5065
> Hi
>
> It is pretty well known that without “help” ( beyond the filter stuff on the super) the 5065 has fairly poor temperature
> stability and somewhat variable response to pressure and humidity. Taking two samples right off the production line
> likely would have given you different long term data as a result.
>
> If you are going to the trouble of watching these devices for a long time, monitor temperature, pressure and humidity.
> I suspect you will find correlations ….
>
> Bob
>
> > On Jun 5, 2018, at 5:44 AM, timeok at timeok.it wrote:
> >
> >
> > Hi Attila,
> > I am doing the long therm stability test on a couple of HP5065, one original A type, the second B or "super" compared with my best long therm stability, a GPSDO HPz3816A.
> > This test will be 30 days long. It is a second test, with a second couple of HP5065A/B, to verify the long therm stability of both the version.
> >
> > Up to now the data available for version B or "super" only concerned the improvement of phase noise.
> >
> > From the first results it seems that the HP5065A is more stable within 30days than the modified version with the filter.
> > It must be considered that the short-term stability of 5065 is better than the GPSDO / TICC up to 10-80k sec then, in this test configuration the valid data will be beyond these values.
> > The results will be ready within 28 days and I will be happy to share them in order to have everyone's opinion.
> > Regards,
> > Luciano
> >
> >
> > Da "time-nuts" time-nuts-bounces at febo.com
> > A "Discussion of precise time and frequency measurement" time-nuts at febo.com
> > Cc
> > Data Mon, 4 Jun 2018 10:11:50 +0200
> > Oggetto [time-nuts] Raw phase data of super-5065
> > Hoi zäme!
> >
> > Would one (preferably multiple) of the owners of a super-5065 be so kind
> > and share the raw phase data of an super-5065 against a stable reference
> > with me? The longer the data trace, the better.
> >
> > Thanks in advance!
> >
> > Attila Kinali
> >
> > --
> > It is upon moral qualities that a society is ultimately founded. All
> > the prosperity and technological sophistication in the world is of no
> > use without that foundation.
> > -- Miss Matheson, The Diamond Age, Neil Stephenson
> > _______________________________________________
> > time-nuts mailing list -- time-nuts at febo.com
> > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> > and follow the instructions there.
> > _______________________________________________
> > time-nuts mailing list -- time-nuts at febo.com
> > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> > and follow the instructions there.
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
More information about the Time-nuts_lists.febo.com
mailing list