[time-nuts] Tektronix FCA3103 ADEV measurement tau setting problem
Magnus Danielson
magnus at rubidium.dyndns.org
Tue May 29 20:51:18 UTC 2018
Hej Björn,
On 05/29/2018 10:03 PM, Björn wrote:
> If this is a badged CNT-91...
>
> This is from the current spec.
>
> ”The CNT-91 features a frequency range of 0.001Hz to 400 MHz standard with options to 3, 8, 14 and 20 GHz. ”
>
> The counter was designed while Pendulum still did R&D in Stockholm - not very far from Magnus domains. Magnus certainly knows how to do proper ADEV with that counter.
Yeah, I know the designers. They borrowed some of my counters at one
time for one research project.
Anyway, I have a CNT-90 (++, it's modified in several ways).
I have not really worked with the CNT-91, but don't think it is very
much different than the CNT-90. I have only seen the FCA-3000/3100 at fairs.
So, given that, it should be simple enough to set it up in time-interval
mode, let the start-channel set the pace of the measurement as usual and
then record into TimeLab.
I have not seen any built-in ADEV, that used to be done in a separate
software they had, and see, TimeView is available from Tektronix too as
TVA3000.
TvB has a CNT-91, he can let you know if there is any specifics for that
model to care about.
Cheers,
Magnus
> —
>
> Björn
>
> Sent from my iPhone
>
>> On 29 May 2018, at 20:58, gandalfg8--- via time-nuts <time-nuts at febo.com> wrote:
>>
>>
>> I have a Tek FCA3103 300 MHz counter that measures ADEV
>>
>> as a built in function. When I bring up the settings
>> menu for the measurement, it has an entry window for
>> "tau" (the averaging time, IE "sigma sub y of tau").
>> It defaults to 200 ms. I can enter larger values and
>> ADEV gives reasonably results. However, if I enter
>> smaller values, even 199 ms, I don't get any error
>> on the display, but I get clearly erroneous
>> results for ADEV. I read the manual and cannot find
>> anything to the effect that the instrument doesn't work
>> for less than 200 ms.
>>
>> BTW, I asked Tek "customer no support" about it and
>> they were clueless.
>> -------------------------------
>>
>>
>>
>> Hi Rick,
>>
>>
>> The Tek FCA3100 is a rebadged Pendulum CNT91 and Pendulum became part of Spectracom.
>> Unless I'm missing something, the FCA3103 is the FCA3100 with 3GHz input C fitted.
>>
>>
>> I've also found that Tektronix tech support seemed to know very little about these but, in the past anyway, found Spectracom in the UK to be very helpful. I'm not sure about the current status as Spectracom now seems to redirect to Pendulum as a separate company again.
>>
>>
>>
>> Unfortunately, I can't help with your enquiry, and am also somewhat confused, as I don't recall seeing a settings entry window for "tau" as a settings option for the built in ADEV function. I take it you do mean the built in option as in displaying results on the built in screen, or are you also using some external software?
>>
>>
>> If you are just usin the built in display I'd be interested to hear what firmware version your FCA3103 is showing in the "About" screen
>>
>>
>>
>> My FCA3100 is showing the firmware version as 1.28s of 25 Aug 2010, which seems to match the latest available Tektronix download but I'm wondering now if there are unlisted updates.
>>
>>
>>
>> Nigel, GM8PZR
>>
>>
>>
>>
>>
>>
>>
>>
>>
>>
>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts at febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
More information about the Time-nuts_lists.febo.com
mailing list