[time-nuts] PRS-10 Missing SP values in Appendix A.
Bob kb8tq
kb8tq at n1k.org
Sat Sep 14 02:11:56 UTC 2019
Hi
In a normal PLL the discrete components on the PCB play a big factor
in how the loop behaves. There are some chips that allow the current out
of the detector to be varied a bit, even on those the discrete R’s and C’s are
still the “big deal” in how the loop behaves. There will be a significant impact if
the loop output is 1/3 of what the design was intended to use. In general, that
impact will not be a good one.
Bob
> On Sep 13, 2019, at 7:01 PM, Forrest Christian (List Account) <lists at packetflux.com> wrote:
>
> I suspect someone figured out some of the side effects what you're
> describing and adjusted the divisors to better match across all rows. I
> just wish SRS would have updated the manual if it was them.
>
> In my device, the R,N,A values were 6114,3456,29. The R,N,A values in the
> table in the docs on row 52 are 2038,1145,31. For the discussion below,
> I'm just going to focus on R since the others are effectively related.
>
> The R values 'around' row 52 are 6400,4219,6257,2038,5971,3933 and 5828
> (rows 49 through 55). It seems logical to me to use 6114 as the R
> divisor for row 52 instead of 2038 because the resulting frequency is
> closer to the rest of the values in the table. The highest value I can
> quickly see is 7258 and the lowest is 1180. That is quite a range in
> output frequencies to deal with. If you take the lower values and
> multiply them by a integer multiple so that everything is within a narrower
> range then I'd expect it would be easier to deal with the side effects of a
> now more consistent divided frequency while still permitting the finer
> resolution afforded by a larger divider value. If this is the case, I'd
> also expect the 1180 'lowest value' to have been multiplied by either 4 or
> 5 instead of just 3, and the other lower values in the table to have been
> multiplied by an appropriate range to end up closer to 6000.
>
> Just to add another piece to the big-picture puzzle, the following is from
> the manual:
>
> "The gain of U400’s phase detector may be set (coarsely) by the CPU, and it
> is adjusted to maintain roughly the same PLL damping factor as divisors are
> changed. This loop has a very low natural frequency (about 10 r/s) and a
> damping factor which ranges from 0.84 to 1.19."
>
> U400 is the same MC145190 which is doing the division. So it sounds like
> they're compensating for some of the effects by configuring the phase
> detector differently depending on the divisor values.
>
>
> On Fri, Sep 13, 2019 at 11:00 AM Bob kb8tq <kb8tq at n1k.org> wrote:
>
>> Hi
>>
>> Ummm ….. errrr ……
>>
>> The divisors run down to an output port. There has to be a filter at that
>> port to
>> knock down the noise and make the loop close properly. When you multiply
>> the
>> divisors by three, you cut the frequency at that port by three. You also
>> have a
>> significant impact on the control loop…..
>>
>> Probably a good idea to make sure your board has the “normal” components on
>> it that correspond to the numbers in the manual. There may have been a
>> running
>> change at some point that is not reflected in the doc’s.
>>
>> Bob
>>
>>> On Sep 13, 2019, at 12:12 AM, Forrest Christian (List Account) <
>> lists at packetflux.com> wrote:
>>>
>>> Mainly wanting to post this to the list so it will end up in the
>> archives.
>>>
>>> So I decided to give calibration/adjustment of my bench PRS-10 a go.
>>>
>>> What I discovered is that on my particular unit, the frequency was enough
>>> off that in order to bring it close to spec, I had to adjust the Mag
>> Offset
>>> to the lower end of it's range (2300), and even then the set frequency
>> was
>>> lower than I would have liked.
>>>
>>> According to the section of the manual under the SP command, if the Mag
>>> Offset is at the end of it's range, you can change the frequency
>>> synthesizer's parameters by querying the existing SP? settings, finding
>> the
>>> row in Appendix A which corresponds to those values, then changing the SP
>>> value up or down a step (by using the values in the table row just above
>> or
>>> below the value).
>>>
>>> In my unit, the SP value set (6114,3436,29) were not anywhere to be found
>>> in Appendix A.
>>>
>>> After some digging, and reading the manual, I discovered that these
>> values
>>> are used to configure the MC145193 inside the PRS-10. Specifically the
>>> first value (R) is used to divide the 10Mhz output. The second two
>> values
>>> (N, A) are used to divide 359.72Mhz (which is related to the Rb
>>> frequency). This second divisor is calculated by (N*64+A). The
>> resulting
>>> two divided down signals will be very close in frequency, and the
>>> difference is used to stabilize the oscillator.
>>>
>>> After some more work, I discovered that the divisor values currently in
>> my
>>> oscillator were actually exactly 3 times the value of row 52, that is
>>> 6114/3=2038 and (3436*64+29)/3=(2038*64+31). Since it's only the ratio
>>> between the divisors which matter, I'm assuming someone at some point
>>> decided to use the higher division ratio for some reason. Not sure if
>> this
>>> was at SRS or in the field.
>>>
>>> After discovering this, I followed the procedure to move the SP values by
>>> one row in the table, and everything seems to have re-centered itself.
>>>
>>> Hope this helps someone... Even if it's me in the future if I have to do
>>> this again.
>>>
>>> --
>>> - Forrest
>>> _______________________________________________
>>> time-nuts mailing list -- time-nuts at lists.febo.com
>>> To unsubscribe, go to
>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
>>> and follow the instructions there.
>>
>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts at lists.febo.com
>> To unsubscribe, go to
>> http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
>> and follow the instructions there.
>>
>
>
> --