[time-nuts] OCXO Advice
Tom Van Baak
tvb at LeapSecond.com
Sun Jan 12 16:26:40 UTC 2020
Bob Q,
Both of your UCT 108663 oscillators look ok, although the blue plot
seems 2x to 3x worse than the red plot. Neither is strongly correlated
with temperature. It appears that you're using a GPSDO as a testbed so
some of the noise and strange wiggles seen in your plot may be coming
from the GPSDO board and not the oscillator itself. In that case a
better OCXO might not help.
Each vertical grid line of your plot is 0.5 mHz / 10 MHz = 5e-11. Your
red plot matches mine pretty well:
http://leapsecond.com/pages/uct108663/uct-01.png
BTW, a California eBay seller for UCT 108663 is:
https://www.ebay.com/itm/202863421344
Suggestions:
1) Swap the OCXO's to see if the noise follows the OCXO or follows the
GPSDO board. I mention this because the 2nd hour of the blue plot looks
like GPSDO wander, not typical OCXO noise.
2) Maybe try each OCXO in an isolated test for a day. That is, remove
the GPSDO from the equation -- place the OCXO in an open but undisturbed
no air flow location, power it from a clean stable power supply, and
float or ground the EFC. Measure with your Rb (LPRO) as the reference.
The results of these baseline tests will suggest if the instability that
you see is due to the OCXO or due to the GPSDO boards you're playing with.
3) Consider using both your phase comparators on one OCXO at the same
time to verify the consistency of your measurement system. For example,
the unusual red glitch at T+6 hours looks more like a measurement issue
than something a OCXO would do.
In general, when you have two GPSDO boards, two OCXO, and two counters,
it's not likely they are all equal. So these multiple tests allow you to
gain confidence in your parts and to pinpoint the weak element(s) in
your setup.
/tvb