Thanks very much for this info Corby. It explains the behaviour of my spare
5061A perfectly. 

It also raises the question of what I could do with the spare tube I have. I
can connect a +3500V supply to the ion pump but that won't do anything about
any gas molecules adsorbed onto the filaments as you described. Is it worth
powering up the filaments to get rid of them or can that be left for some
indefinite time in the future when the tube could be reinstalled in a 5061A?

Morris

-------------------------------------------
Date: Wed, 2 Dec 2020 11:06:22 -0800
From: <cdelect at juno.com>
To: time-nuts at lists.febo.com
Subject: [time-nuts] HP5061 Cesium ion pump question

The gettering in the tube is only for capturing any stray Cesium atoms
that don't get caught in the main gettering patch. If the gettering fails
or gets too loaded up then the Cesium background level will get too high
causing poor SN.

The ion pump is for any gases.

When a tube is off for extended times any gas atoms lingering or leaking
slowly into the tube than happen to impinge on either the mass
Spectrometer filament or the Cs oven filaments get capture by the
filaments. They function as excellent getters!  (this even if the ion
pump is on) 

Now this is not by design but results in the filaments being "loaded"
with the gas atoms.

Then when you turn the tube on the filaments light up and expell a burst
of gas.

This of coarse causes the ion pump current to rise and trips off the
filaments,

Once the ion pump removes the burst the cycle repeats until the filaments
have expelled the trapped gases.

Then the ion pump can handle the load and pump the tube down completely.

Cheers,

Corby


*************************