[time-nuts] Re: Can ADEV of a frequency source be correctly determined using a continuous time-stamping frequency counter?
Richard (Rick) Karlquist
richard at karlquist.com
Wed Nov 10 00:53:55 UTC 2021
Let me just mention that when I worked at the HP Santa Clara
Division counters section, they came out with a "feature"
that they called "continuous count". However, it was limited
to something like 3 MHz. So a 100 MHz counter would only
continuously count signals below 3 MHz.
So you need to verify for what bandwidth your specific counter
model is truly doing continuous count.
Rick N6RK
On 11/9/2021 2:29 PM, Magnus Danielson via time-nuts wrote:
> Hi Erik,
>
> On 2021-11-09 18:26, Erik Kaashoek wrote:
>> As far as I understood the ADEV at a Tau of 1 second is a statement
>> about the amount of variation to be expected over a one second interval.
> Rather, the variation of readings of a frequency estimation done over a
> span over 1 second.
>> It would be nice if we would be able to measure a frequency in an
>> infinite short interval but any frequency measurement takes time.
> Turn out that basic white noise and systematic noise will limit our
> frequency resolution to form a 1/tau limit slope, so infinite short
> interval will bury it well into that noise whatever we do.
>> What if the frequency counter does a complete measurement of a
>> frequency source every second and all the variation within that second
>> is hidden because of the "integration" that happens over the second?
>
> That is what happens, but that is not what the ADEV is about, it's about
> the variations of these measures as we look for a bunch of them. So if
> we now have say 1000 of these frequency estimates, how much variations
> in these can be contributed to the random noise of the source, and to
> analyse that, we need at least a tool like ADEV since standard deviation
> will not even converge for white and flicker phase noise modulation.
>
> What ADEV actually aims to do is to provide a low-frequency spectroscopy
> method at a time when time-interval counters was about the only tool at
> hand, and even those where very rare. We now have a much wider palette
> of tools, but ADEV is relevant for how we measure frequency stability
> and a few other applications.
>
>> This is specially the case with continuous time-stamping counters.
>> They can provide a precise number by applying statistical methods on
>> many measurements done during one second but they can not provide
>> information exactly at the end of a second.
>> Is this kind of statistical measurement over a period of a second
>> still valid for determining the ADEV at the Tau of one second of a
>> frequency source?
> Not for ADEV, but if you use averaging counter you get the result of
> MDEV and for linear regression / least square counter you get the
> response of PDEV. That is the result of various statistical measures and
> then applying the ADEV processing on these frequency estimates. The
> upcoming IEEE Std 1139 revision, which is in approval process now
> include language to reflect that.
>> Or should there be a correction factor depending on the method used in
>> the frequency counter?
>
> Yes, you then need to use the appropriate bias function for ADEV/MDEV
> and ADEV/PDEV to convert between these scales. Knowing the response of
> ADEV, MDEV and PDEV for a particular noise-type which is dominant at the
> tau of interest, you can readily convert between them by forming the
> bias functions.
>
> You may find NIST SP-1065 a useful and handy tool, even if it does not
> cover the more recent work such as PDEV.
>
> https://www.nist.gov/publications/handbook-frequency-stability-analysis
>
>> I tried to read some scientific studies on this subject but I am not
>> smart enough to understand.
>> Hope one of you can provide some information.
>
> It is scattered over a large number of articles, and quite a lot of
> folks get confused. Hopefully the updated IEEE Std 1139 will be of aid
> to you. It also has lots of useful references.
>
> Cheers,
> Magnus
> _______________________________________________
> time-nuts mailing list -- time-nuts at lists.febo.com -- To unsubscribe
> send an email to time-nuts-leave at lists.febo.com
> To unsubscribe, go to and follow the instructions there.
>
More information about the Time-nuts_lists.febo.com