[time-nuts] Re: HP 5065A, no 2nd harmonic.
Bob kb8tq
kb8tq at n1k.org
Sat Nov 27 19:49:25 UTC 2021
Hi
Roughly speaking the OCXO is 40 Hz or 8 ppm at the max end of the tune range.
Thatâs a lot.
If your OCXO is that low, then yes, it is a problem. You either need to repair or replace
the OCXO. If you are replacing, the question becomes - do you upgrade to a 10 MHz
10811? That would involve tracking down one of the assemblies that mate up with the
10811.
Bob
> On Nov 27, 2021, at 12:28 PM, Jared Cabot via time-nuts <time-nuts at lists.febo.com> wrote:
>
> Hi,
> I have performed the test with the 10Mohm resistor.
> With an 0.5V P-P signal from the function generator into A7J1, I get 980mV at A7TP2 (Manual says it should be about 0.9V)
> When I probe the A7 output (Yellow wire) i get a good output as described in the manual too, so it looks like my A7 module is ok.
>
> However, when I plug the RVFR back into A7J1, I get exactly no output from the A7 module at all.
>
> Interestingly, the highest I can get the OCXO to adjust using the coarse adjustment (with the fine adjustment set to 250) is
> 4.99996155 MHz, measured using my GPS referenced 5335A connected directly to P3 on the A10A2 OCXO Frequency Divider/Amplifier Assembly.
>
> If this isn't close enough to exactly 5MHz, could it be causing a problem?
>
>
> Thanks,
> Jared
>
>
> âââââââ Original Message âââââââ
>
> On Saturday, November 27th, 2021 at 01:24, paul swed <paulswedb at gmail.com> wrote:
>
>> Jared
>>
>
>> I took a look at the manual and if you inject 137 hz at .5 Vpp through a 10
>> Mohm R thats in a sheild into A7j1 you will see .9 Vpp at TP2 and some
>> large signal upto 16 Vpp at Yellow out. That would say the A7 module is
>> good. It looks like the 2 ICs are lm741 class opamps if they happened to be
>> bad.
>>
>
>> Regards
>>
>
>> Paul
>>
>
>> WB8TSL
>>
>
>> On Wed, Nov 24, 2021 at 8:36 AM Jared Cabot via time-nuts <
>>
>
>> time-nuts at lists.febo.com> wrote:
>>
>
>>> Hi,
>>>
>
>>> Ok, I managed to get a moment to do a few tests. I'll try to answer the
>>>
>
>>> questions so far, sorry if I miss something out.
>>>
>
>>> For test equipment, I have the following gear of possible use:
>>>
>
>>> - HP 5335A 200MHz Universal Counter with 1.3GHz option
>>> - Leo Bodnar GPS Reference
>>> - HP 34461A Multimeter, just calibrated.
>>> - Tek TDS748C 1GHz Scope
>>> - Multiple HP E361xA Power supplies
>>> - 3325A Function Gen (All options)
>>> - Many other bits of gear too, including a billion TM500 modules and
>>>
>
>>> mainframes etc.
>>>
>
>>> On the 5065A front, I have replaced the resistors in the RVFR with
>>>
>
>>> Mil-spec Vishay RN55, RN60 and RN65 series parts, cleaned the glass
>>>
>
>>> elements and added a 100 deg.C thermostat to prevent over temperature
>>>
>
>>> events.
>>>
>
>>> I have replaced all the wet tantalum capacitors in the entire unit with
>>>
>
>>> United Chemicon KYB 220uF, 63V capacitors (10,000 hour rated), bypassed
>>>
>
>>> with C0G/NP0 0.1uF MLCC ceramic caps, there was no PCB damage, but two tant
>>>
>
>>> caps had just started leaking, luckily not reaching the PCB's.
>>>
>
>>> I added heat sinks to the two diodes on the A2 board and replaced the RIFA
>>>
>
>>> mains filter cap too. The 3 big mains filter capacitors tested well within
>>>
>
>>> spec, with leakage down around 10uA.
>>>
>
>>> - On A8 Pin9, I get 137KHz, 174mV P-P
>>>
>
>>> - A7 TP2 shows me random low amplitude noise
>>>
>
>>> - A3 TP2 I see 5.38VDC
>>>
>
>>>
>
>>> My serial number is 2432A01344 and I have the 10811-60109 OCXO (and no
>>>
>
>>> options fitted, so the chassis is clean of battery corrosion).
>>>
>
>>> Thanks all for the help!
>>>
>
>>> Jared_______________________________________________
>>>
>
>>> time-nuts mailing list -- time-nuts at lists.febo.com -- To unsubscribe send
>>>
>
>>> an email to time-nuts-leave at lists.febo.com
>>>
>
>>> To unsubscribe, go to and follow the instructions there.
>>
>
>> time-nuts mailing list -- time-nuts at lists.febo.com -- To unsubscribe send an email to time-nuts-leave at lists.febo.com
>>
>