[time-nuts] RFTGm-II (KS-24019) XO Fault but GPS to Rb is OK
k1jos at att.net
k1jos at att.net
Sun Oct 31 14:19:46 UTC 2021
I have had a KS-24019 (L108-L109) pair running continuously for several
years. Now have a red LED Fault on the XO unit and the RB unit shows green
LED On-Line status. The NO-GPS LEDs are off on both the -XO and -RB units
but if I disconnect the antenna - they do show Yellow NO_GPS status. Got
out my old XP laptop and ran the Lucent software. The Rb (REF 0) is active
and locked to GPS and working fine. The XO says Unit Failed but the message
code for the first 4 hex nibbles which should show alarm status are all zero
bits so no failure code is evident. The last 2 alarm nibbles are 53
representing the # of hours since failure. The GPS time is correct on the
Rb but is fixed on the XO at 80-01-06 00:00:00. The XO checksum shows PASS.
Based on this the XO GPS is working, locked and sending correct GPS messages
to Rb. The XO CPU seems to be working as well so am I correct this seems
to be a Osc failure? Besides checking voltages (which I'm not sure what
they should be) - any advice how to go about further testing sand repair?
Thanks in advance
Jerry NY2KW
--
This email has been checked for viruses by Avast antivirus software.
https://www.avast.com/antivirus
More information about the Time-nuts_lists.febo.com