After some more work it became possible to measure the coherent noise
floor in a similar way as was described in "A Small Dual Mixer Time
Difference (DMTD) Clock Measuring System" (Small DMTD) by W.J. Rely
Both inputs where connected to separate outputs of a two channel DDS.
This allowed checking the frequency and phase measurement by testing
with various frequency difference and phase offsets. The frequency and
phase measured for all frequency and phase variations as expected and
it was fairly simple to measure the minimum phase step of the DDS (0.006
degrees)
For the coherent noise floor tests the frequency difference was set to zero.
Coherent phase data [1]
Over a roughly 45 minutes period the relative phase slowly varied over a
+2ps/-5ps range, comparable to the Small DMTD.
The measurement is susceptible to temperature variations as the
prototype does not yet have an enclosure. Draft from opening doors had
an observable impact, most easily spotted in the phase data.
Frequency data [2]
The frequency appeared to be stable, no linear trend was observed. At
first sight the frequency noise appeared to be less then the Small DMTD
but the Small DMTD did measure the frequency at a lower tau of 0.1 second.
Frequency stability [3]
Although no long term drift was present the slope of the stability plot
suggests there is still some hidden issue. The frequency stability of
the Small DMTD is clearly better.
Frequency histogram [4]
The histogram also hints to some none Gaussian behavior of the noise.
The total width of the histogram seems comparable to the Small DMTD
Auto correlation of the frequency [5]
The frequency auto correlation clearly shows there is some repetitive
signal hidden in the frequency data.
The frequency of the signal, about 1/120 Hz, suggests this could be due
to the 0.01 Hz resolution of the digital PLL that locks the internal
reference to one of the inputs. As the PLL is frozen during measurements
the frequency difference between input and internal reference could
easily be 1/120 Hz
Erik.
[1] http://athome.kaashoek.com/time-nuts/DMTD/phase.PNG
[2] http://athome.kaashoek.com/time-nuts/DMTD/freq.PNG
[3] http://athome.kaashoek.com/time-nuts/DMTD/stability.PNG
[4] http://athome.kaashoek.com/time-nuts/DMTD/histo.PNG
[5] http://athome.kaashoek.com/time-nuts/DMTD/autoc.PNG