[time-nuts] Re: CTI OSC5A2B02 OXCO testing
Tom Van Baak
tvb at LeapSecond.com
Sun Jun 4 16:29:46 UTC 2023
Reg,
Thanks for posting your 10 points of raw data. That will help debug the
problem. I made two plots for you.
Your first point (at 1.5819 V) doesn't look correct so I ignored that.
The 8th point (at 23708 V) looks wrong -- 24 kV is kind of high for an
OCXO -- so I changed it to 2.3708 guessing it was just a transcription
error.
Plot reg-1.png -- For EFC testing of a good oscillator the points tend
to fall very consistently along a path and the path is often a line or a
slightly curved line. See [1] for an old example. But your data is quite
scattered. That might indicate that either your voltage source is very
unstable or your oscillator is unstable or even your frequency counter
is unstable. Or something as mundane as loose wires or noise or trigger
error.
I would suggest before investigating EFC further that you strive to get
clean readings. Perhaps set the EFC to a fixed value, maybe use a 1.5 V
battery, and then collect 10 frequency readings in a row. They should
all agree to 0.1 or or 0.01 Hz. If so then your counter is not the
problem. BTW, what is your counter gate time set to?
Plot reg-2.png -- For the 2nd plot I sorted your data and connected the
dots. Again this should be linear or approximately linear. You can see
the reading (1.7740 10.00000050) is completely out of place suggesting a
problem with your setup. And the 2 or 3 points at the top end don't look
right either.
/tvb
[1] http://leapsecond.com/pages/tcxo-efc/
On 6/3/2023 1:40 PM, Reginald Beardsley via time-nuts wrote:
>
> Here are the results of my preliminary test using just a bench supply and 5386A on the internal XOCO. I just adjusted the Vref until I got within 1 Hz. PSU is settable in 10 mV steps. Data are from my 34401A.
>
> Vref MHz
> 1.5819 10.000000??
> 1.7758 10.00000015
> 1.6924 10.00000010
> 1.7740 10.00000050
> 2.0853 10.00000047
> 1.9374 10.00000040
> 2.1520 10.00000075
> 23708 10.00000073
> 1.9942 10.00000046
> 2.6832 10.00000065
>
> _______________________________________________
> time-nuts mailing list -- time-nuts at lists.febo.com
> To unsubscribe send an email to time-nuts-leave at lists.febo.com
>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: reg-1.png
Type: image/png
Size: 5334 bytes
Desc: not available
URL: <http://febo.com/pipermail/time-nuts_lists.febo.com/attachments/20230604/23290840/attachment.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: reg-2.png
Type: image/png
Size: 5675 bytes
Desc: not available
URL: <http://febo.com/pipermail/time-nuts_lists.febo.com/attachments/20230604/23290840/attachment-0001.png>
More information about the Time-nuts_lists.febo.com
mailing list