[time-nuts] HP 5061A oscillator problem
vilgotch1 at gmail.com
vilgotch1 at gmail.com
Thu Sep 28 22:21:19 UTC 2023
Hi all,
Since 2006 my in-house frequency standard has been a HP 5061A which
still has a good physics package. I have retrofitted a Shera GPS system
to it in the disused battery compartment and run it occasionally on the
Cs to keep it nice.
Recently I turned off all lab power while I was away for a few weeks and
when I turned it back on the 5061A wasn't working properly. It would
start up and run for a few minutes and then the outputs would disappear.
I have traced the problem to the 00105-6013 crystal oscillator unit and
I think what is happening is that the rapid warm up thermostat is stuck
on. When starting from cold it runs OK until the innards get too hot and
then it fails. When I disconnected the AC feed to the rapid warm-up
heater the oscillator runs properly and the osc heater meter starts at
the extreme right end of the scale before slowly backing off over an
hour or so and the oscillator unit feels warm.
What I want to know is:
1. Is this a reasonable explanation and has anyone seen such a fault
before?
2. How long does the 00105-6013 take to get to operating temp using it's
internal controller without the rapid warm-up?
3. Is the crystal oscillator unit repairable? I suspect a replacement
thermostat would be unobtainable.
4. Does anyone have a replacement 00105-06013 unit available? To my
eternal regret I had a spare and sold it a couple of years ago. It just
goes to show you should never throw anything out!
Thanks for your help and 73 to all,
Morris VK3DOC
Melbourne, Australia