[volt-nuts] 3458A - calibration constants coding
Poul-Henning Kamp
phk at phk.freebsd.dk
Sat Nov 5 11:34:01 UTC 2011
In message <4EB50F9C.7000602 at freenet.de>, Frank Stellmach writes:
>The content of the first two constants by MEMREAD:
Which are the exact addresses you did MREAD from ?
You have understood that MREAD reads 16 bits so you should only
give it even addresses and that for the CALRAM only 8 of the 16
bits are valid ?
ID?
HP3458A
MREAD 393216
16569
MREAD 393218
-7239
MREAD 393220
-30791
Which means that the first three bytes are 0x40, 0xe3 and 0x87
>>> "%x" % 16569
'40b9'
>>> "%x" % (65536-7239)
'e3b9'
>>> "%x" % (65536-30791)
'87b9'
>>>
>This is my question: Is the content of the NVRAM, if read out on a
>programmer, also encrypted as described above, or does this occur only
>by the MEMREAD function?
There is no encryption or anything of the sort.
--
Poul-Henning Kamp | UNIX since Zilog Zeus 3.20
phk at FreeBSD.ORG | TCP/IP since RFC 956
FreeBSD committer | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.
More information about the volt-nuts
mailing list