Joe,
...Correction to my previous statement regarding HP 3457A ACAL:
Being so used to the HP 3458A's ACAL functions, I think I incorrectly
attributed some 3458A ACAL features to the 3457A. Although the 3458A ACAL
re-generates new "temporary" digital correction constants for DCV gain,
apparently the 3457A ACAL does *not* do that.
According to the 3457A Service Manual, its ACAL routine adjusts DC offsets
of its AC amplifier, adjusts its AC attenuator flatness, and adjusts its
ohms precharge amplifier.
So, apparently the only time 3457A DCV gain is adjusted is during actual
"real" calibration against external voltage standards.
This being the case, I assume your 3457A's 10Vdc reading will *not* change
immediately after ACAL ALL.
Let us know if that's the case?
-Greg
----- Original Message -----
From: "J. L. Trantham" <jltran at att.net>
To: "'Discussion of precise voltage measurement'" <volt-nuts at febo.com>
Sent: Friday, September 02, 2011 8:28 PM
Subject: Re: [volt-nuts] Solartron 7081, other meters,and some voltage
standards
I have not yet read the manual. Can you educate me about the 'ACAL ALL'?
I presume that the 'calibration' standards loaded at the last official
calibration are stored and can be 'reloaded' when needed. Correct?
Thanks,
Joe
-----Original Message-----
From: volt-nuts-bounces at febo.com [mailto:volt-nuts-bounces at febo.com] On
Behalf Of gbusg
Sent: Friday, September 02, 2011 6:52 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Solartron 7081, other meters,and some voltage
standards
Hi Joe,
.........
After several hours warm-up, be sure to run "ACAL ALL".
..........