[volt-nuts] Reference Calibration Options - Solartron 7081 andFluke 731B and 335A
gbusg
gbusg at comcast.net
Sat Sep 10 01:09:00 UTC 2011
Hi Joe,
Another thing to ask AMTEK is whether or not their 7081 calibration always
includes adjustment? Or do they only adjust if a parameter reaches a
guardband limit or spec limit?
Obviously for your purposes it would be nice if they adjusted the box back
to nominal. But that's not a typical service for most T&M calibrations.
The neat thing about the "golden" calibration (Loveland Standards Lab Cal)
for Agilent's 3458A is that it *always* includes adjustment of the box back
to nominal. But that turns out to be more feasible in the case of the 3458A
because there are only 3 adjustments for the total DCV, DCA and ohms
functions: CAL 0, CAL 10 and CAL 10k. (Note: This *always adjust" policy
only applies to the "golden" 3458A cal, not for example the "silver" cal.)
Let us know what you find out from AMTEK regarding their 7081 cal.
Thanks,
Greg
----- Original Message -----
From: "J. L. Trantham" <jltran at att.net>
To: <volt-nuts at febo.com>
Sent: Wednesday, September 07, 2011 8:58 PM
Subject: [volt-nuts] Reference Calibration Options - Solartron 7081 andFluke
731B and 335A
I have explored the options for calibration of my 'reference' devices.
The Solartron 7081 calibration is available in Oak Ridge, TN, at Advanced
Measurement Technology http://www.ametek-online.com/ for $800.00. They are
the US branch of the company that owns Solartron, if I understanad
correctly. I have emailed with a number of questions regarding the specific
details of the calibration and documents supplied but have not received a
reply so far. I suspect the Service Coordinator I emailed is busy chasing
down the answers for my many questions.
For the 731B, Fluke offers a Performance Test for $229, Calibration,
Traceable, W/O Data for $245, Calibration, Z540 Traceable, W/O Data for $259
and Calibration, Z540 Traceable, W/ Data for $333. There is no repair
service available.
For the 335A, there is no Performance Test option and there, also, is no
repair service available. However, for the remainder, it is $473, $505, and
$662 respectively.
My thought is to get the 7081 calibrated and one of the 731B's. The 'white
face' 731B is the later unit with the modifications hand written in the
manual downloaded from the Fluke website already in place. The 'black face'
unit is older and does not have these modifications installed. I'm thinking
the 'white face' unit would be the best to get calibrated.
Once these are done, I can then do the 'calibration' of the 335A and the
other 731B myself.
Any thoughts about this plan, which calibration to choose, or other options
that might be available? Also, any thoughts about the potential to need
'repair' of these Fluke units when sent for calibration?
Thanks in advance.
Joe
_______________________________________________
volt-nuts mailing list -- volt-nuts at febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
and follow the instructions there.
More information about the volt-nuts
mailing list