Hi Warren,
Here's some data at 15 minute increments.
Notes:
1. Far right column delta calculations are *relative* to the reading at 240 minutes (Reference).
2. The first delta is at 15 minutes from 3458A cold start, thus hints at the 3458A warm-up characteristics.
3. The last reading was taken at 1440 minutes (therefore at a much greater interval than 15 minutes from previous reading).
4. The 10Vdc source is a Fluke 732A. The meter is a HP 3458A.
Also note: I'm sending this message as *Rich Text*, hoping the table survives the group's email process.
If it doesn't survive intact, I'll have to convert the table to jpg or something and re-send as an attachment. (Anyone here have a easy way to convert selected portions of xls files to jpg?)
Proto #1
Readings Readings
Time Before ACAL DC After ACAL DC Delta From Final 240 minute Reading for "Readings After ACAL DC"
0 min 10.0001136 V
15 min 10.0000269 V 10.0000183 V -1.6E-06 V -0.16 ppm
30 min 10.0000180 V 10.0000206 V 7E-07 V 0.07 ppm
45 min 10.0000202 V 10.0000211 V 1.2E-06 V 0.12 ppm
60 min 10.0000204 V 10.0000210 V 1.1E-06 V 0.11 ppm
75 min 10.0000201 V 10.0000209 V 1E-06 V 0.10 ppm
90 min 10.0000199 V 10.0000203 V 4E-07 V 0.04 ppm
105 min 10.0000203 V 10.0000204 V 5E-07 V 0.05 ppm
120 min 10.0000195 V 10.0000202 V 3E-07 V 0.03 ppm
135 min 10.0000196 V 10.0000203 V 4E-07 V 0.04 ppm
150 min 10.0000198 V 10.0000196 V -3E-07 V -0.03 ppm
165 min 10.0000204 V 10.0000207 V 8E-07 V 0.08 ppm
180 min 10.0000204 V 10.0000197 V -2E-07 V -0.02 ppm
195 min 10.0000196 V 10.0000198 V -1E-07 V -0.01 ppm
210 min 10.0000199 V 10.0000198 V -1E-07 V -0.01 ppm
225 min 10.0000199 V 10.0000198 V -1E-07 V -0.01 ppm
240 min 10.0000199 V 10.0000199 V Reference Reference
1440 min 10.0000209 V 10.0000201 V 2E-07 V 0.02 ppm
-Greg
----- Original Message -----
From: "ws at Yahoo" <warrensjmail-one at yahoo.com>
To: <volt-nuts at febo.com>
Sent: Monday, September 26, 2011 10:33 PM
Subject: Re: [volt-nuts] Fluke 335A
Greg
What PP varations do you see over an hour or two?
ws
*************
Hi Joe,
Yes, that's correct.
Greg
----- Original Message -----
From: "J. L. Trantham" <jltran at att.net>
To: "'Discussion of precise voltage measurement'" <volt-nuts at febo.com>
Sent: Monday, September 26, 2011 10:12 PM
Subject: Re: [volt-nuts] Fluke 335A
Greg,
If my math is correct, then you see about +/- 0.3 uV at 10 V. Correct?
Joe
-----Original Message-----
From: volt-nuts-bounces at febo.com [mailto:volt-nuts-bounces at febo.com]
On
Behalf Of gbusg
Sent: Monday, September 26, 2011 10:45 PM
To: Discussion of precise voltage measurement
Subject: Re: [volt-nuts] Fluke 335A
Hi Joe,
The Fluke 732A (at 10Vdc) is by far the most "quiet" source I have here.
With HP 3458A integration time set for 100 Power Line Cycles, (and for the
combination of [Noise Produced by 732A]+[Noise Induced by 3458A]), I'm
observing +/- 0.03ppm short-term noise (racking-around of the 3458A from
reading to reading) at 10Vdc.
Regarding short-term noise of other possible DCV sources and combinations
(including your D cell battery suggestion, homebrew zener references, etc.),
I defer to the expertise and experience of the other fellows in the group,
for their comments. (Fellows, what say regarding Joe's question?)
Best,
Greg