[volt-nuts] Agilent 3458A Positive Experience
Marv Gozum @ JHN
marvin.gozum at jefferson.edu
Fri Feb 10 19:14:55 UTC 2012
This is great to know Joe. Just curious, 'home cal'd' 3458a sounds
like a challenge. When you say 2 of the AC ranges were off, which
ones? Do you have a variable AC voltage and frequency source to use,
or did you use some other substitute?
With the Agilent unit DC gain needing adjustment, did the any DC or
ohms functions read off specifications before the Agilent was calibrated?
At 09:09 02/10/2012, J. L. Trantham wrote:
>requiring me to 're-cal' the meter using 'in house'
>equipment.
>
>After leaving the meters on for several weeks, with all functions seemingly
>correct, I sent both to Agilent for their Agilent Calibration for $579.81
>each. Both passed (with my 'home-cal'd' unit being off on only 2 of the AC
>ranges and the Agilent unit being off only in the DC Gain) and were returned
>in about 10 days, amazingly fast.
Best Wishes,
Marv Gozum
Philadelphia
More information about the volt-nuts
mailing list