[volt-nuts] Thermistor value for 732A Reference 10V Standard
Rob Klein
rob.klein at smalldesign.nl
Thu Jun 28 23:45:21 UTC 2012
Hi Tom,
Definitely not good! The thermistor is supposed to be an NTC, so seeing
the value
rise from cold means there's something very wrong indeed.
If you want to replace it, this might be a good alterative:
http://www.digikey.nl/product-detail/en/192-103LET-A01/480-3158-ND/1886044
Kind regards,
Rob Klein.
Op 29-6-2012 1:04, Leedyt at aol.com schreef:
>
> Hi!
>
> Can anyone give me a rough value of the thermistor resistance that is
> brought out to the front panel of a Fluke 732A DC Reference Standard (lower
> left-hand side)? The thermistor is for the user to monitor the temperature of
> the oven that contains the voltage reference. I see a resistance that
> starts out at 6600 Ohms (unit was nearly cold) and climbs and levels out at
> about 8400 Ohms after a day, or so. Is this reasonable? The thermistor is
> given in the parts list as RT2, a Fenwal JA41J1 (no longer made) and on the
> schematic as a 10K @ 25C unit. So this doesn't make sense.
>
> Does anyone have any experience with this? Or is just simpler to replace
> the thermistor with a 100 Ohm RTD and be done with it? Or would this drive
> the calibration labs nuts?
>
> Any references or advice would be appreciated.
>
> Tom Leedy
> Clarksburg, MD
> _______________________________________________
> volt-nuts mailing list -- volt-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
More information about the volt-nuts
mailing list