[volt-nuts] HP3458A calibration memory backup
Poul-Henning Kamp
phk at phk.freebsd.dk
Thu Aug 7 10:23:49 EDT 2014
--------
In message <012401cfb249$f429ca90$dc7d5fb0$@att.net>, "J. L. Trantham" writes:
>I would love to know where the CALNUM value is stored.
This is a dump of the CALRAM, with device-addresse. If you use
MREAD, you need to multiply the address by two and add 0x60000:
CalRam:
00060000 40e387b0724e1ca5 .double 3.999751395e+04 @0x000 40K reference
00060008 401caec53f49b86c .double 7.170674313e+00 @0x008 7V reference
00060010 3f5f2bea74341701 .double 1.902560195e-03 @0x010 dcv zero front 100mV
00060018 bf6c3c70cfa013b1 .double -3.446789106e-03 @0x018 dcv zero rear 100mV
00060020 3f2926695a613d43 .double 1.918796162e-04 @0x020 dcv zero front 1V
00060028 bf36b0763e594f67 .double -3.462113076e-04 @0x028 dcv zero rear 1V
00060030 3efcb169c5e30e8c .double 2.736379644e-05 @0x030 dcv zero front 10V
00060038 bf01c710cb49a0bc .double -3.390808107e-05 @0x038 dcv zero rear 10V
00060040 bf374e66fc1c76ca .double -3.556252887e-04 @0x040 dcv zero front 100V
00060048 bf374e66fc1c76ca .double -3.556252887e-04 @0x048 dcv zero rear 100V
00060050 bf00cd21f93b4ade .double -3.204593591e-05 @0x050 dcv zero front 1KV
00060058 bf00cd21f93b4ade .double -3.204593591e-05 @0x058 dcv zero rear 1KV
00060060 402bb6444b89982a .double 1.385598980e+01 @0x060 ohm zero front 10
00060068 3ff633a34dd87bce .double 1.387606911e+00 @0x068 ohm zero front 100
00060070 3fc1bde269b326f1 .double 1.386073128e-01 @0x070 ohm zero front 1K
00060078 3f8cc0105f68f191 .double 1.403820792e-02 @0x078 ohm zero front 10K
00060080 3f47a45e5a6c9462 .double 7.214985570e-04 @0x080 ohm zero front 100K
00060088 3f1a1c078c6ab664 .double 9.959980080e-05 @0x088 ohm zero front 1M
00060090 3f0380e3b5ff8529 .double 3.719992560e-05 @0x090 ohm zero front 10M
00060098 3f09cb7fbf28a7db .double 4.919990160e-05 @0x098 ohm zero front 100M
000600a0 3f09cb7fbf28a7db .double 4.919990160e-05 @0x0a0 ohm zero front 1G
000600a8 402a64ce170cea26 .double 1.319688484e+01 @0x0a8 ohm zero rear 10
000600b0 3ff50df9128fc784 .double 1.315911362e+00 @0x0b0 ohm zero rear 100
000600b8 3fc0d58a2688d03c .double 1.315167130e-01 @0x0b8 ohm zero rear 1K
000600c0 3f8a5e569146a74d .double 1.287524825e-02 @0x0c0 ohm zero rear 10K
000600c8 3f449f46a7f26f40 .double 6.293387413e-04 @0x0c8 ohm zero rear 100K
000600d0 3f05138ab849cdd5 .double 4.019991960e-05 @0x0d0 ohm zero rear 1M
000600d8 bef4c302eb07bf4d .double -1.979996040e-05 @0x0d8 ohm zero rear 10M
000600e0 bef92a7024a48ac9 .double -2.399995200e-05 @0x0e0 ohm zero rear 100M
000600e8 bef92a7024a48ac9 .double -2.399995200e-05 @0x0e8 ohm zero rear 1G
000600f0 3f82b38a38bf8728 .double 9.131507737e-03 @0x0f0 ohmf zero front 10
000600f8 3f83033c9e21a6a4 .double 9.283517433e-03 @0x0f8 ohmf zero front 100
00060100 3f4f2e3b28813062 .double 9.515560969e-04 @0x100 ohmf zero front 1K
00060108 3f50039aedf92f74 .double 9.774220452e-04 @0x108 ohmf zero front 10K
00060110 3f1a0bec635d8049 .double 9.935980128e-05 @0x110 ohmf zero front 100K
00060118 3f1bff365bf70d9a .double 1.067997864e-04 @0x118 ohmf zero front 1M
00060120 3f19a33bd887a097 .double 9.779980440e-05 @0x120 ohmf zero front 10M
00060128 3f1b866aa813f7cc .double 1.049997900e-04 @0x128 ohmf zero front 100M
00060130 3f1b866aa813f7cc .double 1.049997900e-04 @0x130 ohmf zero front 1G
00060138 3f7ba0d2627aed67 .double 6.745168510e-03 @0x138 ohmf zero rear 10
00060140 3f7c266bad8003b7 .double 6.872578255e-03 @0x140 ohmf zero rear 100
00060148 3f47d7ceb2552c63 .double 7.276305447e-04 @0x148 ohmf zero rear 1K
00060150 3f4667aeb672694e .double 6.837466325e-04 @0x150 ohmf zero rear 10K
00060158 3f11c9ff975782bf .double 6.785986428e-05 @0x158 ohmf zero rear 100K
00060160 3f1216809a5643c1 .double 6.899986200e-05 @0x160 ohmf zero rear 1M
00060168 3f097af7f1e69952 .double 4.859990280e-05 @0x168 ohmf zero rear 10M
00060170 3f0b5e26c172f088 .double 5.219989560e-05 @0x170 ohmf zero rear 100M
00060178 3f0b5e26c172f088 .double 5.219989560e-05 @0x178 ohmf zero rear 1G
00060180 0000019f .long 415 @0x180 autorange offset ohm 10
00060184 00000029 .long 41 @0x184 autorange offset ohm 100
00060188 00000004 .long 4 @0x188 autorange offset ohm 1k
0006018c 00000000 .long 0 @0x18c autorange offset ohm 10k
00060190 00000000 .long 0 @0x190 autorange offset ohm 100k
00060194 00000000 .long 0 @0x194 autorange offset ohm 1M
00060198 00000000 .long 0 @0x198 autorange offset ohm 10M
0006019c 00000000 .long 0 @0x19c autorange offset ohm 100M
000601a0 00000000 .long 0 @0x1a0 autorange offset ohm 1G
000601a4 40414b85974038ab .double 3.459001437e+01 @0x1a4 cal 0 temperature
000601ac 4041514611239ab1 .double 3.463495077e+01 @0x1ac cal 10 temperature
000601b4 404149398d69e0ae .double 3.457206886e+01 @0x1b4 cal 10k temperature
000601bc aaaf .word 43695 @0x1bc Cal_Sum0
000601be b6 .byte -74 @0x1be vos dac
000601bf > -- -- -- -- -- -- -- -- -- -- -- -- -- -- -- a0 _______________
000601c0 bf426fca5f844e1c .double -5.626428747e-04 @0x1c0 dci zero rear 100nA
000601c8 bf78af31e9d15976 .double -6.026453947e-03 @0x1c8 dci zero rear 1uA
000601d0 bf7822864b92dc98 .double -5.892300215e-03 @0x1d0 dci zero rear 10uA
000601d8 bf77b55affb88373 .double -5.788188424e-03 @0x1d8 dci zero rear 100uA
000601e0 bf77472c64fc6f34 .double -5.683110634e-03 @0x1e0 dci zero rear 1mA
000601e8 bf770b99887177c5 .double -5.626296747e-03 @0x1e8 dci zero rear 10mA
000601f0 bf41c815f7445785 .double -5.426509147e-04 @0x1f0 dci zero rear 100mA
000601f8 bf4105b493067f51 .double -5.194789610e-04 @0x1f8 dci zero rear 1A
00060200 3f33693d1af3e1af .double 2.961897071e-04 @0x200 dcv gain 100mV
00060208 3f6848aa8440ff6f .double 2.964337387e-03 @0x208 dcv gain 1V
00060210 3f9e55f2377d8995 .double 2.962473358e-02 @0x210 dcv gain 10V
00060218 3fd2f88eb57edac0 .double 2.964207432e-01 @0x218 dcv gain 100V
00060220 4007b2e12408fc1b .double 2.962343484e+00 @0x220 dcv gain 1KV
00060228 3f9eb11e485fea1d .double 2.997252766e-02 @0x228 ohm gain 10
00060230 3fd32e157a157817 .double 2.996877377e-01 @0x230 ohm gain 100
00060238 4007fea962909739 .double 2.999346514e+00 @0x238 ohm gain 1K
00060240 403dfeed478ccdf9 .double 2.999580810e+01 @0x240 ohm gain 10K
00060248 4082bd915e4ac5c6 .double 5.996959806e+02 @0x248 ohm gain 100K
00060250 40b7691fc3ba945a .double 5.993124080e+03 @0x250 ohm gain 1M
00060258 40ed46779435036b .double 5.995573684e+04 @0x258 ohm gain 10M
00060260 40ed46779435036b .double 5.995573684e+04 @0x260 ohm gain 100M
00060268 40ed46779435036b .double 5.995573684e+04 @0x268 ohm gain 1G
00060270 3f9eb11e4c3e9af0 .double 2.997252789e-02 @0x270 ohm ocomp gain 10
00060278 3fd32e159243bc53 .double 2.996877602e-01 @0x278 ohm ocomp gain 100
00060280 4007fea980d0cc93 .double 2.999346739e+00 @0x280 ohm ocomp gain 1k
00060288 403dfeeeb4fbb2f7 .double 2.999582988e+01 @0x288 ohm ocomp gain 10k
00060290 4082bd9336fc618f .double 5.996968822e+02 @0x290 ohm ocomp gain 100k
00060298 40b7691fc3ba945a .double 5.993124080e+03 @0x298 ohm ocomp gain 1M
000602a0 40ed46779435036b .double 5.995573684e+04 @0x2a0 ohm ocomp gain 10M
000602a8 40ed46779435036b .double 5.995573684e+04 @0x2a8 ohm ocomp gain 100M
000602b0 40ed46779435036b .double 5.995573684e+04 @0x2b0 ohm ocomp gain 1G
000602b8 3e37552cee544b95 .double 5.432571481e-09 @0x2b8 dci gain 100nA
000602c0 3e3c1ae6a6502707 .double 6.543724317e-09 @0x2c0 dci gain 1uA
000602c8 3e6e32d9db309d0a .double 5.624934420e-08 @0x2c8 dci gain 10uA
000602d0 3e9b21aa3291b60d .double 4.042909035e-07 @0x2d0 dci gain 100uA
000602d8 3ec8d48f0a8f4c88 .double 2.960003333e-06 @0x2d8 dci gain 1mA
000602e0 3efec119e6ca646b .double 2.932958744e-05 @0x2e0 dci gain 10mA
000602e8 3f665d4b0b8ce9d7 .double 2.730032513e-03 @0x2e8 dci gain 100mA
000602f0 3fa0a3295d7b4910 .double 3.249482409e-02 @0x2f0 dci gain 1A
000602f8 45 .byte 69 @0x2f8 precharge dac
000602f9 55 .byte 85 @0x2f9 mc dac
000602fa 4014097d30835a7b .double 5.009266623e+00 @0x2fa high speed gain
00060302 3dd46a24a2c83445 .double 7.426796044e-11 @0x302 il
0006030a bd860b055295ca6f .double -2.506004644e-12 @0x30a il2
00060312 4163131689fcef6c .double 1.000056431e+07 @0x312 rin
0006031a bf7d3f146331a7bb .double -7.140235556e-03 @0x31a low aperture
00060322 bfafad3829b42f60 .double -6.186843404e-02 @0x322 high aperture
0006032a 3feffffd454fd517 .double 9.999986986e-01 @0x32a high aperture slope .01 PLC
00060332 3fefffff8f2da21f .double 9.999997899e-01 @0x332 high aperture slope .1 PLC
0006033a 409a0bf4ea37aece .double 1.666989175e+03 @0x33a high aperture null .01 PLC
00060342 40d046be0ecdf649 .double 1.666696965e+04 @0x342 high aperture null .1 PLC
0006034a 000013c8 .long 5064 @0x34a underload dcv 100mV
0006034e 000013c4 .long 5060 @0x34e underload dcv 1V
00060352 000013c7 .long 5063 @0x352 underload dcv 10V
00060356 000013c4 .long 5060 @0x356 underload dcv 100V
0006035a 000013c7 .long 5063 @0x35a underload dcv 1000V
0006035e 0000ed60 .long 60768 @0x35e overload dcv 100mV
00060362 0000ed30 .long 60720 @0x362 overload dcv 1V
00060366 0000ed54 .long 60756 @0x366 overload dcv 10V
0006036a 0000ed30 .long 60720 @0x36a overload dcv 100V
0006036e 0000ed54 .long 60756 @0x36e overtoad dcv 1000V
00060372 0000138c .long 5004 @0x372 underload ohm 10
00060376 0000138d .long 5005 @0x376 underload ohm 100
0006037a 00001389 .long 5001 @0x37a underload ohm 1k
0006037e 00001388 .long 5000 @0x37e underload ohm 10k
00060382 000009c5 .long 2501 @0x382 underload ohm 100k
00060386 000009c6 .long 2502 @0x386 underload ohm 1M
0006038a 000009c5 .long 2501 @0x38a underload ohm 10M
0006038e 000030dd .long 12509 @0x38e underload ohm 100M
00060392 000058d8 .long 22744 @0x392 underload ohm 1G
00060396 0000ea90 .long 60048 @0x396 overload ohm 10
0006039a 0000ea9c .long 60060 @0x39a overload ohm 100
0006039e 0000ea6c .long 60012 @0x39e overload ohm 1k
000603a2 0000ea60 .long 60000 @0x3a2 overload ohm 10k
000603a6 0000753c .long 30012 @0x3a6 overload ohm 100k
000603aa 00007548 .long 30024 @0x3aa overload ohm 1M
000603ae 0000753c .long 30012 @0x3ae overload ohm 10M
000603b2 00024a5c .long 150108 @0x3b2 overload ohm 100M
000603b6 00042a20 .long 272928 @0x3b6 overload ohm 1G
000603ba 0000138c .long 5004 @0x3ba underload ohm ocomp 10
000603be 0000138d .long 5005 @0x3be underload ohm ocomp 100
000603c2 00001389 .long 5001 @0x3c2 underload ohm ocomp 1k
000603c6 00001388 .long 5000 @0x3c6 underload ohm ocomp 10k
000603ca 000009c5 .long 2501 @0x3ca underload ohm ocomp 100k
000603ce 000009c6 .long 2502 @0x3ce underload ohm ocomp 1M
000603d2 000009c5 .long 2501 @0x3d2 underload ohm ocomp 10M
000603d6 000030dd .long 12509 @0x3d6 underload ohm ocomp 100M
000603da 000058d8 .long 22744 @0x3da underload ohm ocomp 1G
000603de 0000ea90 .long 60048 @0x3de overload ohm ocomp 10
000603e2 0000ea9c .long 60060 @0x3e2 overload ohm ocomp 100
000603e6 0000ea6c .long 60012 @0x3e6 overload ohm ocomp 1k
000603ea 0000ea60 .long 60000 @0x3ea overload ohm ocomp 10k
000603ee 0000753c .long 30012 @0x3ee overload ohm ocomp 100k
000603f2 00007548 .long 30024 @0x3f2 overload ohm ocomp 1M
000603f6 0000753c .long 30012 @0x3f6 overload ohm ocomp 10M
000603fa 00024a5c .long 150108 @0x3fa overload ohm ocomp 100M
000603fe 00042a20 .long 272928 @0x3fe overload ohm ocomp 1G
00060402 00000114 .long 276 @0x402 underload dci 100nA
00060406 0000010a .long 266 @0x406 Cal_406
0006040a 00000a6a .long 2666 @0x40a Cal_40a
0006040e 00000e7e .long 3710 @0x40e Cal_40e
00060412 000013cb .long 5067 @0x412 Cal_412
00060416 000013fa .long 5114 @0x416 Cal_416
0006041a 00000225 .long 549 @0x41a Cal_41a
0006041e 000001cd .long 461 @0x41e Cal_41e
00060422 00000cf0 .long 3312 @0x422 overload dci 100nA
00060426 00000c78 .long 3192 @0x426 Cal_426
0006042a 00007cf8 .long 31992 @0x42a Cal_42a
0006042e 0000ade8 .long 44520 @0x42e Cal_42e
00060432 0000ed84 .long 60804 @0x432 Cal_432
00060436 0000efb8 .long 61368 @0x436 Cal_436
0006043a 000019bc .long 6588 @0x43a Cal_43a
0006043e 0000159c .long 5532 @0x43e Cal_43e
00060442 40440527ae21b385 .double 4.004027344e+01 @0x442 acal dcv temperature
0006044a 40440575371cb708 .double 4.004263963e+01 @0x44a acal ohm temperature
00060452 404404c54531e04c .double 4.003727021e+01 @0x452 acal acv temperature
0006045a 82 .byte -126 @0x45a ac offset dac 10mV
0006045b 83 .byte -125 @0x45b ac offset dac 100mV
0006045c 82 .byte -126 @0x45c ac offset dac 1V
0006045d 81 .byte -127 @0x45d ac offset dac 10V
0006045e 83 .byte -125 @0x45e ac offset dac 100V
0006045f 83 .byte -125 @0x45f ac offset dac 1KV
00060460 82 .byte -126 @0x460 acdc offset dac 10mV
00060461 83 .byte -125 @0x461 acdc offset dac 100mV
00060462 82 .byte -126 @0x462 acdc offset dac 1V
00060463 81 .byte -127 @0x463 acdc offset dac 10V
00060464 83 .byte -125 @0x464 acdc offset dac 100V
00060465 83 .byte -125 @0x465 acdc offset dac 1KV
00060466 89 .byte -119 @0x466 acdci offset dac 100uA
00060467 89 .byte -119 @0x467 acdci offset dac 1mA
00060468 89 .byte -119 @0x468 acdci offset dac 10mA
00060469 89 .byte -119 @0x469 acdci offset dac 100mA
0006046a 89 .byte -119 @0x46a acdci offset dac 1A
0006046b > -- -- -- -- -- -- -- -- -- -- -- a0 -- -- -- -- ___________ ____
0006046c 09f3 .word 2547 @0x46c flatness dac 10mV
0006046e 09ee .word 2542 @0x46e flatness dac 100mV
00060470 05a5 .word 1445 @0x470 flatness dac 1V
00060472 0a70 .word 2672 @0x472 flatness dac 10V
00060474 0b48 .word 2888 @0x474 flatness dac 100V
00060476 0b4d .word 2893 @0x476 flatness dac 1KV
00060478 80 .byte -128 @0x478 level dac dc 1.2V
00060479 > -- -- -- -- -- -- -- -- -- a0 -- -- -- -- -- -- _________ ______
0006047a 7e .byte 126 @0x47a level dac dc 12V
0006047b > -- -- -- -- -- -- -- -- -- -- -- a0 -- -- -- -- ___________ ____
0006047c 7d .byte 125 @0x47c level dac ac 1.2V
0006047d 7e .byte 126 @0x47d level dac ac 12V
0006047e 7f .byte 127 @0x47e dcv trigger offset 100mV
0006047f 7e .byte 126 @0x47f dcv trigger offset 1V
00060480 7e .byte 126 @0x480 dcv trigger offset 10V
00060481 7e .byte 126 @0x481 dcv trigger offset 100V
00060482 7e .byte 126 @0x482 dcv trigger offset 1000V
00060483 > -- -- -- a0 -- -- -- -- -- -- -- -- -- -- -- -- ___ ____________
00060484 c08c0ed532617c1c .double -8.978541000e+02 @0x484 acdcv sync offset 10mV
0006048c c08e5772bd3c3611 .double -9.709310250e+02 @0x48c acdcv sync offset 100mV
00060494 c08e6da64c2f837b .double -9.737062000e+02 @0x494 acdcv sync offset 1V
0006049c c08e2f05532617c2 .double -9.658776000e+02 @0x49c acdcv sync offset 10V
000604a4 c08e7ce189374bc7 .double -9.756101250e+02 @0x4a4 acdcv sync offset 100V
000604ac c08e676db8bac711 .double -9.729285750e+02 @0x4ac acdcv sync offset 1KV
000604b4 c08c15a4816f0069 .double -8.987053250e+02 @0x4b4 acv sync offset 10mV
000604bc c08e59dab9f559b4 .double -9.712318000e+02 @0x4bc acv sync offset 100mV
000604c4 c08e825367a0f909 .double -9.762907250e+02 @0x4c4 acv sync offset 1V
000604cc c08e34573eab367a .double -9.665426000e+02 @0x4cc acv sync offset 10V
000604d4 c08e8479c0ebedfa .double -9.765594500e+02 @0x4d4 acv sync offset 100V
000604dc c08e70bed288ce70 .double -9.740931750e+02 @0x4dc acv sync offset 1KV
000604e4 3ebfcf93873671dd .double 1.896074107e-06 @0x4e4 acv sync gain 10mV
000604ec 3ef49bb8893272f9 .double 1.965359168e-05 @0x4ec acv sync gain 100mV
000604f4 3f2a8369e0822d9b .double 2.022806877e-04 @0x4f4 acv sync gain 1V
000604fc 3f5fef5e2724d425 .double 1.949159571e-03 @0x4fc acv sync gain 10V
00060504 3f93b85a1296b4bb .double 1.925793398e-02 @0x504 acv sync gain 100V
0006050c 3fc8a2e4dedfae90 .double 1.924711312e-01 @0x50c acv sync gain 1KV
00060514 40d46f8232f22aab .double 2.092603436e+04 @0x514 ab ratio
0006051c 4024bac2bba7bf38 .double 1.036476695e+01 @0x51c gain ratio
00060524 3efe9d73bdb291b4 .double 2.919678432e-05 @0x524 acv ana gain 10mV
0006052c 3f33d566e134877d .double 3.026367351e-04 @0x52c acv ana gain 100mV
00060534 3f698444cd4d07e9 .double 3.114828471e-03 @0x534 acv ana gain 1V
0006053c 3f9ebc0c6d7ddbdb .double 3.001422327e-02 @0x53c acv ana gain 10V
00060544 3fd2fa9472c77460 .double 2.965441819e-01 @0x544 acv ana gain 100V
0006054c 4007b5cff657c656 .double 2.963775563e+00 @0x54c acv ana gain 1KV
00060554 3f99b6bf8096d9b8 .double 2.511119101e-02 @0x554 acv ana offset 10mV
0006055c 3f99b6bf8096d9b8 .double 2.511119101e-02 @0x55c acv ana offset 100mV
00060564 3f99b6bf8096d9b8 .double 2.511119101e-02 @0x564 acv ana offset 1V
0006056c 3f99b6bf8096d9b8 .double 2.511119101e-02 @0x56c acv ana offset 10V
00060574 3f99b6bf8096d9b8 .double 2.511119101e-02 @0x574 acv ana offset 100V
0006057c 3f99b6bf8096d9b8 .double 2.511119101e-02 @0x57c acv ana offset 1KV
00060584 3ff06e1846b0689c .double 1.026878620e+00 @0x584 rmsdc ratio
0006058c 3ff0013b8286a9e5 .double 1.000300894e+00 @0x58c sampdc ratio
00060594 3feffd60540f89f6 .double 9.996797220e-01 @0x594 aci gain
0006059c b241 .word 45633 @0x59c Cal_Sum1
0006059e 3fb9956bbd47c715 .double 9.993623134e-02 @0x59e Cal_59e
000605a6 3feffaf7c66cf761 .double 9.993857265e-01 @0x5a6 Cal_5a6
000605ae 400806557a9ef6a0 .double 3.003092726e+00 @0x5ae Cal_5ae
000605b6 3ff0001100066002 .double 1.000016213e+00 @0x5b6 freq gain
000605be 3d .byte 61 @0x5be attenuator high frequency dac
000605bf > -- -- -- -- -- -- -- -- -- -- -- -- -- -- -- a0 _______________
000605c0 3e .byte 62 @0x5c0 amplifier high frequency dac 10mV
000605c1 3e .byte 62 @0x5c1 amplifier high frequency dac 100mV
000605c2 3e .byte 62 @0x5c2 amplifier high frequency dac 1V
000605c3 4e .byte 78 @0x5c3 amplifier high frequency dac 10V
000605c4 50 .byte 80 @0x5c4 amplifier high frequency dac 100V
000605c5 50 .byte 80 @0x5c5 amplifier high frequency dac 1KV
000605c6 09 .byte 9 @0x5c6 interpolator
000605c7 > -- -- -- -- -- -- -- a0 -- -- -- -- -- -- -- -- _______ ________
000605c8 10bb .word 4283 @0x5c8 Cal_Sum2
000605ca a0a0a0a0a0a0a0a0a0a0a0a0... .asciiC " "
0006061a 0000004d .long 77 @0x61a Calnum
0006061e 00000d82 .long 3458 @0x61e Cal_SecureCode
00060622 00 .byte 0 @0x622 Cal_AcalSecure
00060623 > -- -- -- a0 -- -- -- -- -- -- -- -- -- -- -- -- ___ ____________
00060624 337f .word 13183 @0x624 Cal_Sum3
00060626 0000001b .long 27 @0x626 Destructive Overloads
0006062a 00000000 .long 0 @0x62a Defeats
0006062e .fill 0xa0 [0xea]
00060718 .fill 0x1 [0x30]
00060748 > -- -- -- -- -- -- -- -- 11 11 11 11 11 10 10 70 ________ p
00060750 .fill 0x1 [0x27]
00060777 > -- -- -- -- -- -- -- 07 04 04 04 04 04 04 04 04 _______
00060780 > 04 44 44 44 44 44 44 44 44 44 44 44 44 44 44 44 DDDDDDDDDDDDDDD
00060790 > 44 44 44 44 44 44 44 44 44 44 44 44 44 44 44 44 DDDDDDDDDDDDDDDD
000607a0 > 04 04 04 fc ff 57 55 f5 17 -- -- -- -- -- -- -- WU _______
000607a9 .fill 0x10 [0x22]
000607cb > -- -- -- -- -- -- -- -- -- -- -- 70 40 40 40 40 ___________p@@@@
000607d0 > 7f 01 00 00 00 00 00 00 00 00 00 44 f4 a4 -- -- D __
000607de .fill 0xa0 [0x21]
--------------------------------------------------
--
Poul-Henning Kamp | UNIX since Zilog Zeus 3.20
phk at FreeBSD.ORG | TCP/IP since RFC 956
FreeBSD committer | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.
More information about the volt-nuts
mailing list