A few things to keep in mind:
- the 720A (and the 752A) are self-calibrating, i.e. you can (easily) calibrate it yourself before use. The 3458A, using its external artifact self-cal procedure based on only 10v and 10k, requires a performance verification therafter (at least every second time, see some military docs concluding this, I don't recall the link but easy to find; i would actually say, to comply with GUM, every time).
-the linearity of the 3458a is excellent and not beatable up to a 1:10 ratio, beyond that (e.g. if you want to calibrate an instruments 1000v or 100mV range form a 10V reference), it is not usable, again the 720a or 752A are superior.
So to summarize, if you want to be selfstanding, and do your calibrations with reference to an externally calibarted 10v and 10k resistor only, you need a good voltage source (5440 is hard to beat even by the 5720A, and can replace, with some compromises, also a 732a if need be), a 3458A but also some type of self-cal 1:10 and 1:100 divider (720A or 752 or similar). You could build a hammon divider (1:10 and 1:100) yourself, for much less than a 720A costs, and since it is self-cal, if you do it right, you do not need to compromise on accuracy. what you really need is just the decade divide ratios to do all the 3458A validation, and then go from there. all the gazillion other ratios that the 720a offers are really not needed then.
Adrian
> Gesendet: Donnerstag, 23. April 2015 um 21:19 Uhr
> Von: "Frank Stellmach" <frank.stellmach at freenet.de>
> An: volt-nuts at febo.com
> Betreff: [volt-nuts] Advise to Junior Member Regarding Acquisition of Fluke 5XXX Series Calibrator
>
> Hi Stan,
>
> the 720A can easily be replaced by an HP3458A, which is superior to the
> KV divider in several aspects.
>
> First, the 3458A has 3-10 times better linearity, 0.02ppm (typ.) of F.S.
> compared to 0.1ppm of input for the 720A.
> Therefore, a 10:1 transfer is accurate to 1ppm for the 720A only,
> whereas the 3458A manages 0.1 to 0.3 ppm.
>
> The self calibration is much easier on the 3458A, as are all these
> calibration measurements, you mentioned.
> (Fluke has published a good application note, how to replace their own
> old style equipment as KV, Null VM, etc. by their own 8 1/2 digit 8508A).
>
> And you may get a very reasonable and recent instruments for 3000$/€,
> maybe much less for older ones.
>
>
> A calibrator is limited in use, as it needs a Null VM at least.
>
> Anyhow, I recommend the Fluke 5440A / 5442A DCV calibrators.
>
> They are ultra stable, having 2 stacked SZ263A references inside, good
> for 732A stability.
>
> Their D/A is also extremely linear, I measured something like 0.2ppm INL
> against my 3458A, and they are spec'd to 0.5ppm of output, also superior
> to the 720A in some volt areas.
>
> They also have this handy autocal function (like the 3458A) for the 4
> higher volt ranges ( 11, 22, 220, 1000V), once that they are externally
> calibrated, and because their internal component drift is low, after
> these years.
> That means, as their internal reference is very stable, you may bring it
> near 24hr. specification, every time you do the autocal.
> I could not measure any deviation to that, using my 3458A, plus a self
> built precision Hammon divider for 1kV.
>
> Their autocal feature is not explicitely promoted, but you will find
> that description "between the lines" in the addendum of the user manual.
>
> Only the 1V and 100mV range have to be externally calibrated quite often.
>
> These instruments may cost about 2000$/€.
>
> So you better have both, the 5440 and the 3458A.
>
> Frank
> _______________________________________________
> volt-nuts mailing list -- volt-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts
> and follow the instructions there.
>
More information about the volt-nuts
mailing list