In reply to Li Ang, Bruce wrote:
>The CLK1 input circuit produces an output incompatible with the 3.3V
>CMOS device it drives.A pair of pnp transistors in an otherwise
>similar circuit is capable of producing a 3.3V CMOS compatible output signal.
>Using independent voltage dividers to bias the transistor bases is a
>bad idea in that resistor tolerances may lead to a dc input offset
>of several tens of millivolts even with 1% resistors,
I agree with Bruce. The circuit below avoids these problems. It is
generally known as a "Wenzel squarer" (after Charles Wenzel, who
popularized it -- see
<http://www.wenzel.com/library/time-frequency-articles/waveform-conversion-part-i-sine-to-square/>).
I revised and simulated a circuit I use all the time for 5v output to
produce a 3v output, but I did not build it, so some adjustment may
be required. All resistors should be metal film. The 1uF capacitors
should be X7R, and the 100nF and 10nF capacitors should be
NP0/C0G. "Design center" for this circuit is a 10MHz input at 1Vrms
(not shown is the 50 ohm input resistor that would terminate a 1Vrms,
50 ohm source).
Note that the circuit needs some "overhead" voltage to bias the PNP
devices, so a 5v power supply is shown. Both this supply and the
base reference supply need to be quieter than the precision you
expect from the circuit. The decoupling shown should be
significantly better than required for your purposes here, but keep
this in mind if you push on to significantly higher resolution.
I don't know what the highest frequency you expect to count is. The
3906s are as fast as the 3904s you are using, and are fine at 10MHz
and even up toward 100MHz -- but at some point you would need faster
transistors. The MMBT5179 and BFT93 are two possibilities. Note
that these faster transistors also reduce the sloping of the top of
the "square" wave output (which is due to capacitive feedthrough of
the B-E junction of the input transistor). With less current
available from this feedthrough (due to lower junction capacitance),
you will probably need to increase the output resistor (R6) to
achieve a full 3v output level if you use faster transistors. [Note
-- the attached simulated scope traces show the output (Q2 collector)
and reference (Q2 base). You want to keep Q2 out of saturation, so
the peak collector voltage needs to be no more positive than the base
voltage, as shown.]
Note also that the input capacitor and the emitter coupling capacitor
limit the lowest frequency you can count.
Finally, note that the circuit does not have a lot of gain, so the
low-signal limit is higher than on most commercial counters. A
0.2Vp-p input produces a 0-2.6v, mostly sinusoidal output. By 1Vp-p
input, it is nicely square. I'd limit the input voltage to
~5Vp-p. If you need better sensitivity, you can add a preamp. (If
this were to be used as a general-purpose counter, I'd design a
limiting preamp for the input.)
Best regards,
Charles
-------------- next part --------------
A non-text attachment was scrubbed...
Name: Wenzel_squarer_3v_AC_2N3906.gif
Type: image/gif
Size: 92191 bytes
Desc: not available
URL: <http://febo.com/pipermail/time-nuts_lists.febo.com/attachments/20141225/47dd95be/attachment.gif>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: Wenzel_3v_squarer_scope_traces_sm.png
Type: image/png
Size: 53257 bytes
Desc: not available
URL: <http://febo.com/pipermail/time-nuts_lists.febo.com/attachments/20141225/47dd95be/attachment.png>
More information about the Time-nuts_lists.febo.com
mailing list