[time-nuts] Lucent KS-24361 GPSDO arrived today several questions
Bob Camp
kb8tq at n1k.org
Thu Nov 6 12:04:43 UTC 2014
Hi
> On Nov 6, 2014, at 3:44 AM, GandalfG8--- via time-nuts <time-nuts at febo.com> wrote:
>
> Hi Paul
>
> Prime question is what are you calling the "slave" unit?
1) Slave in the sense that it’s slaved to the GPS in the other box. Unless we find a serial link from the Ref-0 back into the Ref-1, that’s all it’s doing.
2) It fires up and the Ref-1 box gets all organized and locked up. It goes “ok” first, then the Ref-0 goes “ok” (locks to GPS). In that sense, the Ref-1 is the master.
My guess is that they default to output from the Ref-0 rather than Ref-1 because it’s the one more likely to fail (more single points to go wrong).
> In normal use Lucent treats the Ref-0 unit as the master and Ref-1 as the
> slave, but it's Ref-1 that contains the GPS module. Bit confusing
Yes it is confusing - sorry about that
> but if
> you've got Ref-1 with the GPS module then you're in business:-)
>
> It's turned out the transistor delay circuit isn't needed after all and the
> wiring can be simplified too.
>
> All that's necessary to run the Ref-1 stand alone is to ground pins 2 and 3
> on the J5 interface connector and there's a couple of convenient ground
> pins already on that connector, 8 and 13.
> In normal use pin3 is tied hard to ground, presumably via direct link from
> the other module, so that can just be a wire link to either pin 8 or 13.
> Pin 2 is normally held at a logic low level so although a direct ground
> seems to work ok for that too, and no damage noted so far, I prefer to play
> it safe and ground that one to the other ground pin via a 470ohm resistor,
> which also works fine.
> That's it.
So a single 15 pin connector with one jumper and one resistor (or jumper) plugged into the interface socket.
>
> The LEDS should flash at start up, followed by "No GPS" and "Fault" on
> solid whilst the unit goes through acquisition and lock, this can take an hour
> or so, and then those LEDs should go off and the "On" light should be
> solid.
> At this point the 15MHz and 1PPS outputs are enabled.
>
> The J8 diagnostics connector is enabled from start up and SatStat can be
> used to monitor progress.
> The RS422 3 wire fudge suggested by Stu Cobb works fine for me, J8 pin 7 to
> PC 9 way comm port pin 5, J8 pin 8 to comm port pin 3, J8 pin 9 to comm
> port pin 2.
Works, but if you have an old style +/- 12 V RS-232 signal, things might or might not work forever. A lot depends on how they have protected the chips. They expect a 1.5 to 3.5V signal on the input, so even +/-5 is a bit exciting. Full blown RS-422 interfaces are sub $15 gizmos.
Bob
>
> Regards
>
> Nigel
> GM8PZR
>
>
>
> In a message dated 06/11/2014 01:24:33 GMT Standard Time,
> paulswedb at gmail.com writes:
>
> I have been following the threads on the KS-24361 some of the longest I
> have ever seen on time nuts.
> So I have a very basic question.
> There was a 2 transistor delay circuit by Arthur way back. I believe this
> is needed.
> But then the pin numbers seem to get a bit mixed up and there are some
> jumpers with several comments.
> Is there a clear guidance on what to do? I do not have the slave unit.
> Power is not an issue.
> Thanks in advance for your help.
> Must say the units very nice.
>
> Regards
> Paul
> WB8TSL
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
More information about the Time-nuts_lists.febo.com
mailing list